Kodak Dryview Cr850 Cassette Problem - Service Technicians Forum

Kodak Dryview CR850 cassette problem

(244)
(33)

CRAIG WAMBURA

Kodak Dryview CR850 cassette problem

January 14, 2020 10:29

I need help with a Kodak Dryview CR850. One applications start-up it displays error 10036-00000, Cassette detected.....

But there is no cassette inserted.

Can anyone help me with resolving this issue please

Log inor Register

to rate and post a comment

(39)
(9)

Suarez Luis

re: Kodak Dryview CR850 cassette problem

January 16, 2020 10:27

10036 - DR_ENTRY_BLOCKED_ON_STARTUP (Classic/Elite)
Description:
During startup, a CASSETTE was detected by the ENTRYSENSOR.
Cause 1:
A CASSETTE is in the INPUT SLOT.
Action:
Remove the CASSETTE.
Cause 2:
The ENTRY SENSOR malfunctioned.
Action:
Install a new ENTRY SENSOR.

Log inor Register

to rate and post a comment

(244)
(33)

CRAIG WAMBURA

re: re: Kodak Dryview CR850 cassette problem

January 16, 2020 05:40

thanks Luis

i will replace the entry sensor

Log inor Register

to rate and post a comment