KodakCarestream Classic Cr Scanner Error 955 -12003 - Service Technicians Forum

Kodak/Carestream Classic CR "Scanner Error" 955 -12003"

Tags: PACS / DR / CR / Digital, KODAK

(2)

Charit Chilpuri

Kodak/Carestream Classic CR "Scanner Error" 955 -12003"

December 05, 2019 10:39

Hi,
Cassette cannot be loaded. The image on the plate may still be valid and will not be erased. Please continue to eject the cassette.
Can anyone kindly help me with this issue?
THANKS.

Log inor Register

to rate and post a comment

(351)
(72)
(2)

juan pablo amado

re: Kodak/Carestream Classic CR Scanner Error 955 -12003

December 06, 2019 09:53

12003 - PLATE PRESENT SENSOR S5 did not detect PLATE
Cause 1: The EXTRACTION BAR SENSOR S9 operates correctly, but another malfunction
occurred.
Action: 1. Remove any obstructions from the path of the EXTRACTION BAR.
2. Do the checkout for the SLOW SCAN CONTROLLER BOARD A6 and COIL
BOARD A7.
Cause 2: The EXTRACTION BAR SENSOR S9 malfunctioned.
Action: Do the checkout for the EXTRACTION BAR SENSOR S9.
Cause 1: The PLATE PRESENT SENSOR S5 operates correctly, but another malfunction
occurred.
Action: 1. Check:
• PLATE was removed from the CASSETTE
• SLOW SCAN MOTOR operates correctly
2. Do the checkout for the SLOW SCAN CONTROLLER BOARD A6 and COIL
BOARD A7.
3. Remove any obstructions from the path of the EXTRACTION BAR.
Cause 2: The PLATE PRESENT SENSOR S5 malfunctioned.
Action: Do the checkout PLATE PRESENT SENSOR S5.