Juan de la Cruz Martinez

Lightspeed characterization

May 03, 2021 12:57

Hi to all, I have a problem with a Lightspeed VCT. The table IMS won't characterize. I have replaced the absolute encoder and the GTCB. The up/down, and in/out characterize without any problem, but the IMS keeps failing the characterization on the third step. Thanks

Log inor Register

to rate and post a comment

(1)

G2DIAZ

re: Lightspeed characterization

May 06, 2021 10:23

Hello Juan. Did you get some table error on the error log