dismiss

Mark your calendars: the next Clean Sweep Live Auction will be on Thursday, June 21st Click to view the full catalogue

SEARCH
Current Location:
>
>
>
> Logiq 200 Ultrasound Screen Brightness Issue

Logiq 200 Ultrasound Screen Brightness Issue

Tags: Ultrasound, GE

(1)
(1)

Laurence Boss

Logiq 200 Ultrasound Screen Brightness Issue

January 31, 2018 09:27

I'm currently trying to fix a LOGIQ 200 Ultrasound in a provincial hospital in Cambodia. The CRT monitor displays at a very high brightness making it unusable and the brightness dial has no effect on the screen brightness.
Any help on this problem would be greatly appreciated.

1 person liked this
Log inor Register

to rate and post a comment

(90)
(8)

Ramon Portillo

re: Logiq 200 Ultrasound Screen Brightness Issue

February 13, 2018 08:08

Hi Laurence, your problem is a resistor that is broken, I mean open. If you write to me I will email the service manual.

Your problem is common in CRT monitors, so check the resistors where the grid or Bias... The problem will be in a voltage divider circuit. Anything you may contact to me at: ramon.portillo.m@gmail.com

Log inor Register

to rate and post a comment

Repairs for less with FREE Transducer EVALS! CALL NOW 1-888-838-7488

If you have a broken or defective probe, why not attempt to repair before purchasing? This could save you $1,000’s! With a free evaluation we can let you know the cost of repair if applicable. Call today: 1-888-838-7488 Email: sales@sentinelimaging.net DOTmed Certified DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED