Elias Valdes

Lorad Affinity

October 10, 2007 09:47

Hello:
I have a problem with a Lorad Affinity mammo unit, evry time an exposure is performed, I get Error 44 CONTROL, followed by Error 45 CONTROL and Error 25 CONTROL. The exposure is very short and the the film is very light. I will appreciatte any ideas, Thanks.
E Valdes

Log inor Register

to rate and post a comment

(331)
(61)
(13)

Manuel Casanova

Re: Lorad Affinity

October 10, 2007 07:39

Elias:

The problem appears in manual mode? or only in automatic mode.

Check your AEC.

best regards

Manuel

Log inor Register

to rate and post a comment

(57)
(6)

Elias Valdes

Re: Lorad Affinity

October 11, 2007 09:44

Thanks, it happens in any mode, manual or auto.

Log inor Register

to rate and post a comment

Ralph Childs

Re: Lorad Affinity

October 12, 2007 09:40

Elias,

Your tube is most likely starting to fail.

Scooter
scanner1@eim.ae

Log inor Register

to rate and post a comment

(57)
(6)

Elias Valdes

Re: Lorad Affinity

October 13, 2007 12:07

Thanks, those are bad news.

Log inor Register

to rate and post a comment

(17)

(1)

Rodrigo Piva

Re: Lorad Affinity

December 11, 2007 09:56

The problem is the arcing tube, the only way to repair is changing the tube


Bad News

Regards
Rodrigo

Log inor Register

to rate and post a comment

Infusion Pump Repair Corp, Certified Expert Repair

Infusion Pump Repair (IPR) is a certified medical support and solutions company. We specialize in the repair, refurbishment and resale of infusion pumps. Call 1-855-477-8866 DOTmed Certified

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED