Lorad M-IV Mammography System Error: 34

January 19, 2017 06:04

Dear All,
Please our Lorad M-IV Mammo displays Error 34. Any idea on what could have caused this and how to resolve it?
I will also appreciate to have the service manual for the Mammo (dophs_ak@yahoo.com).

Thank you.

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(616)
(17)

Lenny Place

re: Lorad M-IV Mammography System Error: 34

January 23, 2017 07:57

34 Displayed Message: System Error 34.
Error: Focal Spot Relay Fault.
Notes: The condition of the focal spot relay on the filament control board does not match the
condition the generator processor has selected.

Check relay and troubleshoot the control signal on the filament board signal comes from the microprocessor board.

Try to reload backups I had this issue once due to corrupt calibration files

Log inor Register

to rate and post a comment

(312)
(22)
(8)

Joseph Jenkins

re: re: Lorad M-IV Mammography System Error: 34

October 16, 2020 10:29

hello ..do you have the service manual / diagrams for this system? I need a copy asap.
getting error 62 (motion) when doing a LMAO using either bucky

Log inor Register

to rate and post a comment

(45)
(8)

carlos enrique salazar gea

re: Lorad M-IV Mammography System Error: 34

January 23, 2017 07:59

System Error 34 mA Control Focal spot selection fault Check the focal spot relay on the
Filament Control board. Replace
board if necessary.

1 person liked this
Log inor Register

to rate and post a comment

(93)
(16)

Juancho

re: Lorad M-IV Mammography System Error: 34

January 23, 2017 08:03

Hi, Dophs. Send me your email. Thanks

Log inor Register

to rate and post a comment

(52)
(3)

Nadeem Hameed

re: re: Lorad M-IV Mammography System Error: 34

January 13, 2020 02:05

please share me details to solve this error

Log inor Register

to rate and post a comment

(312)
(22)
(8)

Joseph Jenkins

re: re: Lorad M-IV Mammography System Error: 34

October 16, 2020 10:31

hello ..do you have the service manual / diagrams for this system? I need a copy asap.
getting error 62 (motion) when doing a LMAO using either bucky

Log inor Register

to rate and post a comment

(52)
(3)

Nadeem Hameed

re: Lorad M-IV Mammography System Error: 34

January 13, 2020 02:05

share me details to solve this error thanks

Log inor Register

to rate and post a comment

Brandon Transfer & Storage - Rigging, Moving & Set Up For Medical Equipment

Family owned since 1919, for your next medical move, rely on the specialists at Brandon Transfer & Storage. For references from the medical community & a free moving estimate, call us at 561-683-5000 and we care! Visit www.brandontransfer.com Today!

advertisement

 

RSTI Newsletter

 

Hitachi

 

one source

Upcoming Auction Events

ZRG LLC.
TIMED
Dec 11, 2020 (45 lots)