Mammo Ge Senographe 800t E12 071081 - Service Technicians Forum
advertisement
Current Location:
>
>
>
> Mammo GE Senographe 800T E12 071/081

Mammo GE Senographe 800T E12 071/081

Tags: Mammo, GE

(106)
(9)

Svilen Popov

Mammo GE Senographe 800T E12 071/081

August 07, 2019 11:20

Hello, I have problems with a GE Senographe 800T, showing the following error: E12 Checksum Error 071/081. I changed the ram-back up battery, but had this error again. Also saw, that led "FPGA OK" on GENERATOR & ARM INTERFACE BOARD - flashes constantly , and haven't movements of the device. What I must to do ??? Any ideas ???

P.S. I haven't backup floppies of the system.

1 person liked this
Log inor Register

to rate and post a comment

(1)

Lucio Moraes

re: Mammo GE Senographe 800T E12 071/081

August 08, 2019 09:19

E12 Checksum Error Generator : Checksum error on Battery Backā€“Up RAM (BBR) 071/081

Log inor Register

to rate and post a comment

(1)

Rich_SRS

re: Mammo GE Senographe 800T E12 071/081

August 08, 2019 09:20

Reset the checksum and proceed to calibrate generator and arm. Can also load databases from another unit to verify everything works (before calibrating for 4~5 hours). Enjoy.

Log inor Register

to rate and post a comment

(25)

(2)

Raul Cervantes

re: Mammo GE Senographe 800T E12 071/081

August 08, 2019 09:21

If You do not have the backup floppy of the system, You need recalibrate all (generator and arm), You need software of service, system manual, cable of conection, lap-top with serial port, and other accesories. and experience in the equipment.

Log inor Register

to rate and post a comment

(106)
(9)

Svilen Popov

re: Mammo GE Senographe 800T E12 071/081

August 08, 2019 09:25

Someone to have back-up copy with calibrated settings and parameters for this unit ?

Log inor Register

to rate and post a comment

(25)

(2)

Raul Cervantes

re: re: Mammo GE Senographe 800T E12 071/081

August 09, 2019 10:01

Is possible but the automatic exposition will not work only the manual. I can send you the backup if you want

Log inor Register

to rate and post a comment

(106)
(9)

Svilen Popov

re: re: re: Mammo GE Senographe 800T E12 071/081

August 11, 2019 02:23

I already begun calibrations and made generators calibration. If you could send me backup file with Arm calibrations, because I haven't this "force detector" to measure compression in daNs, and it will be very difficult without it(to calibrate the Arm)

Log inor Register

to rate and post a comment

Nationwide Service & Support 1-866-NUC-MED1

OEM trained nationwide service for NM, PET, PET/CT systems. Relocation, installation, parts, technical and clinical support. Contact us at info@nis-mit.com

 