Hi, I need a part 2 of the manual, which describes how to configure after replacing the CPU board.
My Alfa ST shows the error E12 and the picture is only visible to the contour of the breast
thank you
good morning i have problem with Alpha st Mammo ,the problem was when you press the paddle for the compression breast doesn't moving in both direction up/down and after checking i found the back transformer is damaged and after replacing it the compression is start going up to max but doesn't moving down ,if you have any idea about that pls replay to me biotechmeds@gmail.com
Seems that C&Z driver board is broken. Check the part of this board which control the compression motor. Disconect motor and test voltages. Maybe transistors are bad an give alwais voltage to motor. Maybe relay is broken (always on)
Thank you friend, are really many patient and the problem only appears in the end of the day, this could be the lack of air conditioning indiciios room temperature brand 18ºc
I have manuals for all Instrumentarium Units. Post your email address. Take the cover off and feel the tube when this happens. Also check the thermo in the tube and see if it is open at the time of error.
That will tell you if the tube is really hot or if is software related.
Hi, you said you have Service Manual for GE Alpha ST Mammograph. Would you send me a copy? I also need the Parts List. Please, let me know if it is possible. My e-mail address is ademir@secrel.com.br.
Cancel that last post, it doesn't have a thermal in the tube, but if you turn the anode brake off that will keep the tube from getting as hot. If there are very busy the brake can be generating extra heat. Send me your email, you turn off the brake in the service program.
BUg, y have the same problem but, E03 come on when the first patient also , E01, when I press the x ray button again, it lets me fire, and some time it does not, I jus cannot figure it out where is the problem
Most probably you need to replace the 2 chips in old board (EEPROM with soft and the memory) to the new one. Seems that chip in new board is configured as alpha RT (needs a c-arm control board) and ST doesn't have, so cpu can't find. Other way is enter in software parameters (without replacing chips) and change type of mamo from RT to ST.
Malik Ibragim
Mammograph Alpha ST
July 22, 2013 10:15
Hi, I need a part 2 of the manual, which describes how to configure after replacing the CPU board.
My Alfa ST shows the error E12 and the picture is only visible to the contour of the breast
thank you
to rate and post a comment
(4)
Douglas Gray
re: Mammograph Alpha ST
July 23, 2013 08:39
E12 is a communication error between the CPU and c-arm control board. You probably need to use the software chip out of the old board.
to rate and post a comment
Mohammed abu al ata
re: re: Mammograph Alpha ST
November 26, 2013 08:17
Dear Douglass,,,
good morning i have problem with Alpha st Mammo ,the problem was when you press the paddle for the compression breast doesn't moving in both direction up/down and after checking i found the back transformer is damaged and after replacing it the compression is start going up to max but doesn't moving down ,if you have any idea about that pls replay to me biotechmeds@gmail.com
thanks in advance
to rate and post a comment
(7)
Jose Antonio Galindo
re: re: re: Mammograph Alpha ST
November 26, 2013 03:19
Seems that C&Z driver board is broken. Check the part of this board which control the compression motor. Disconect motor and test voltages. Maybe transistors are bad an give alwais voltage to motor. Maybe relay is broken (always on)
to rate and post a comment
(3)
aurelio samuel
re: re: Mammograph Alpha ST
September 24, 2014 08:10
error e03, appears at the end of the workday and generally turn off the machine and back to work, could someone help med with this problem. thank you
to rate and post a comment
(4)
Bug
re: re: re: Mammograph Alpha ST
September 24, 2014 05:30
Eo3 is tube getting hot. If
At the end of a busy day this might be normal.
to rate and post a comment
(3)
aurelio samuel
re: re: re: re: Mammograph Alpha ST
September 25, 2014 10:32
Thank you friend, are really many patient and the problem only appears in the end of the day, this could be the lack of air conditioning indiciios room temperature brand 18ºc
to rate and post a comment
(3)
aurelio samuel
re: re: re: re: Mammograph Alpha ST
September 25, 2014 10:32
where can I get the service manual of this device ??
to rate and post a comment
(4)
Bug
re: re: re: re: re: Mammograph Alpha ST
September 26, 2014 10:10
I have manuals for all Instrumentarium Units. Post your email address. Take the cover off and feel the tube when this happens. Also check the thermo in the tube and see if it is open at the time of error.
That will tell you if the tube is really hot or if is software related.
to rate and post a comment
ADEMIR VASCONCELOS
re: re: re: re: re: re: Mammograph Alpha ST
February 05, 2016 08:46
Hi, you said you have Service Manual for GE Alpha ST Mammograph. Would you send me a copy? I also need the Parts List. Please, let me know if it is possible. My e-mail address is ademir@secrel.com.br.
Thank you
Ademir
to rate and post a comment
sophx
re: re: re: re: re: re: Mammograph Alpha ST
February 13, 2018 04:40
Hello. If possible, would you be able to send me the service manual for the GE Alpha ST? Thank you.
My email: isoph96@gmail.com
to rate and post a comment
(4)
Bug
re: re: re: re: re: Mammograph Alpha ST
September 26, 2014 10:10
Cancel that last post, it doesn't have a thermal in the tube, but if you turn the anode brake off that will keep the tube from getting as hot. If there are very busy the brake can be generating extra heat. Send me your email, you turn off the brake in the service program.
to rate and post a comment
(9)
Sunil aggarwal
re: re: re: re: re: re: Mammograph Alpha ST
February 29, 2016 08:58
can i can manual.my email is aggarwal@sk-biomex.com
to rate and post a comment
(2)
CIM
re: re: re: re: Mammograph Alpha ST
April 19, 2016 05:21
BUg, y have the same problem but, E03 come on when the first patient also , E01, when I press the x ray button again, it lets me fire, and some time it does not, I jus cannot figure it out where is the problem
to rate and post a comment
(7)
Jose Antonio Galindo
re: Mammograph Alpha ST
July 25, 2013 05:52
Most probably you need to replace the 2 chips in old board (EEPROM with soft and the memory) to the new one. Seems that chip in new board is configured as alpha RT (needs a c-arm control board) and ST doesn't have, so cpu can't find. Other way is enter in software parameters (without replacing chips) and change type of mamo from RT to ST.
to rate and post a comment