Maquet Servo I Tech Error 27

Tags: Respiratory, MAQUET

(2)

Sonal Chand

Maquet Servo I Tech Error 27

April 20, 2017 10:20

Kindly seeking assistance if someone can help me to resolve a technical error '27' appearing when the servo i ventilator is turned on. Also when trying to do Pre Use Check, the internal test fails.

The service manual says to check/replace PC1772 board which I have already done but it did not solve the error.

Thanks for your help.

Log inor Register

to rate and post a comment

(4)

(44)

Jiansheng Meng

re: Maquet Servo I Tech Error 27

April 21, 2017 10:33

Hello Sonal,

This is Amanda. Regarding your questions, pls let me know your email, I will email you. Here is my email mon-vent@dododo.com.cn

Log inor Register

to rate and post a comment

(18)
(1)

Victor Terechtchenko

re: Maquet Servo I Tech Error 27

April 25, 2017 08:19

Hello. This error 27: Tech_err_buzzer_failure. Replace buzzer from PC-1772 plate and from service change volume down of buzzer

Log inor Register

to rate and post a comment