Need help connecting to a Phillips Digital Spot Imaging Unit
Getting an Error on the console that says Check HCU

Thanks

Log inor Register

to rate and post a comment

(225)
(31)
(27)

Wayne Horsman

re: NAI The Dicom Box model AAm-CA DVI

December 02, 2017 11:53

I'm guessing that it is "Host Control" (meaning the Philips system directly controls the NAI Dicom Box), but we don't use Host Control on the NAI Dicom Boxes we support, so we can't really offer much help. I would recommend cloning the HDD, so that you have a replacement ready when it fails.

Log inor Register

to rate and post a comment

(2)

Mark Preble

re: re: NAI The Dicom Box model AAm-CA DVI

December 05, 2017 05:19

Wayne,
Thank you I found that it pinned inside the box for an RS232 data when it was a RS422.
changed the pins and it worked fine.