SEARCH
Current Location:
>
>
> Need help with Datex Ohmeda S5 Monitor please

Need help with Datex Ohmeda S5 Monitor please

(3)

Mark Creger

Need help with Datex Ohmeda S5 Monitor please

December 12, 2011 04:07

I am having a problem with some Datex Ohmeda S5 Patient Monitors having SRAM errors. Changed out the SRAM battery, but now am getting "Select Config" error code on main screen and the machine is not reading the Pulse Oximeter on screen when I boot them up, even though the pulse ox is lit up on the finger probe. I have what I believe to be the master service code of 16, 4, 34, but am unsure exactly how and where to go to "reset config" in the system.
Thank you for any help you can provide.

Log inor Register

to rate and post a comment

(1)
(1)
(1)

Fausto Echeverria

Re: Need help with Datex Ohmeda S5 Monitor please

December 12, 2011 10:18

this is how u get to the service:
Service Menu
To enter the service menu:
1. Press the Monitor Setup key.
2. Select Install/Service (password
16-4-34).
3. Select Service (password 26-23-8).

1 person liked this
Log inor Register

to rate and post a comment

Paul Raffaelli

Re: Need help with Datex Ohmeda S5 Monitor please

December 13, 2011 10:17

If you still need help, contact us for free tech support 877-470-8013...no strings...

Log inor Register

to rate and post a comment

(3)

Mark Creger

Re: Need help with Datex Ohmeda S5 Monitor please

December 13, 2011 05:48

Thanks, i'll be giving you a call. The "Select Config" error code is gone, but now there is a small "%" symbol where the SPO2 should be displaying. The finger probe is lighting up, but still no joy...

Log inor Register

to rate and post a comment

Gaspar Cruz

re: Re: Need help with Datex Ohmeda S5 Monitor please

February 08, 2017 10:18

Hi Mark, I'm having the same issue with the Monitor of a Datex Ohmeda Anesthesia Machine. I had the SRAM error so I changed the memory's battery and the error message disappeared, however, now I'm getting the following error "Patient Monitor Removed"... I'd appreciate so much if you could tell me how you solved your problem.

Log inor Register

to rate and post a comment

Kenan Rahimli

re: Re: Need help with Datex Ohmeda S5 Monitor please

December 25, 2014 10:52

Hey Paul, thanks for your help, it worked for me too, but the only problem is when i reboot the device, the same error displays. I did try to save the new configuration but i am not sure if i did it right. It would be really nice to get some help on this, awaiting your response buddy, thanks in advace...

Log inor Register

to rate and post a comment

(3)

Mo Malik

re: Re: Need help with Datex Ohmeda S5 Monitor please

June 09, 2014 06:38

please advise how to change name of clinic showing on top left corner of the screen

Log inor Register

to rate and post a comment

almotaz bellah

re: Need help with Datex Ohmeda S5 Monitor please

May 22, 2013 07:54

dear mark ; i have the same problem with my monitors so can you tell me about the SRAM and it's location on the control board . I hope that i could have an answer . thanks

3 people liked this
Log inor Register

to rate and post a comment

(5)
(1)

Baris SOZUAK

re: re: Need help with Datex Ohmeda S5 Monitor please

March 05, 2014 07:51

Hi,
You must remove the lcd then remove the shield which covers main board behin the lcd. On the main board you will find a yellow square part . It is the sram and its backup battery with quartz.

1 person liked this
Log inor Register

to rate and post a comment

(9)
(4)

Rick87

re: re: re: Need help with Datex Ohmeda S5 Monitor please

April 09, 2014 03:30

Mark, did you ever resolve problem? I need to replace SRAM as well and dont want to caue anymore problems. Can you say where you aquired the SRAM part? Thanks

1 person liked this
Log inor Register

to rate and post a comment

MICHAEL BATTEN

re: re: re: re: Need help with Datex Ohmeda S5 Monitor please

June 13, 2014 05:11

Hi Mark, have been following this , and also want to know where to purchase the SRAM component for a S/5 Light, Michael Batten

Log inor Register

to rate and post a comment

(9)
(4)

Rick87

re: re: re: re: re: Need help with Datex Ohmeda S5 Monitor please

December 06, 2017 04:23

Hi , I found the battery at Battery Clinic. 952 884-2933. I put in SRAM battery before and this time it ruined the logic board after it was turned on the next day. Unit powered up fine the day i did the work.

Log inor Register

to rate and post a comment

fra ostos

re: Need help with Datex Ohmeda S5 Monitor please

April 14, 2014 12:11

Hello,

after setting up the appliance with pw 04.16.34 and 26.23.8 must save the new configuration under the service menu within the pw 13:20:31

Log inor Register

to rate and post a comment

Gaspar Cruz

re: Need help with Datex Ohmeda S5 Monitor please

February 08, 2017 10:18

Hi Mark, I did the same (changed the memorys' battery) however I'm having the following message "Patient Monitor Removed". How did you solved your problem? I would appreciate your help!

Log inor Register

to rate and post a comment

(3)

Louis Apostolakis

re: Need help with Datex Ohmeda S5 Monitor please

January 03, 2018 02:04

I have the same error. Machine is working despite the error. I opened the machine and have idenitified the part (the yellow rectangle). Do I simply pull it off? I have found what I believe to be the replacement. M4T32-BR12SH1 Is this correct?

Log inor Register

to rate and post a comment

(3)

Louis Apostolakis

re: Need help with Datex Ohmeda S5 Monitor please

January 03, 2018 02:05

Not sure if this is going to be a double reply, but I don't see my first. In short, I opened the monitor, found the yellow battery/crystal mounted to the board. I believe it is simply popped off. But want to be certain. Also, there is a sticker applied to it with an exclaimation point. To me, that means there is something I should be careful about, but I'm not sure what.

Log inor Register

to rate and post a comment

(3)

Louis Apostolakis

re: re: Need help with Datex Ohmeda S5 Monitor please

January 09, 2018 02:09

So, after no response, I just did it. I pulled off the old one. I had to pull hard, but it eventually popped off. I lined up the new carefully, and pushed hard, and it popped back on. I booted the unit, it still said sram error. I factory reset it, booted again, and now the error is gone. I have rebooted about a half dozen times now and no error. I configured my main page again and saved it, reset the clock, and all looks to be in order.

Log inor Register

to rate and post a comment

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We buy, repair, & sell new, used, and refurbished ultrasound probes and stock systems. 5000+ probes in stock. All probes are thoroughly tested & come with a warranty. Contact us about our special offer probes today! 317-759-9210 or sales@probomedical.com DOTmed Certified DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED