Over 90 Total Lots Up For Auction at One Location - WA 04/08

OEC 9400 Flipper Stuck

Tags: C-Arm, OEC

(4)

Ruben Ortiz

OEC 9400 Flipper Stuck

July 03, 2017 08:05

The C-Arm is presenting this error. The manual talks about the collimator. Someone has some idea of ​​what may be happening.

Log inor Register

to rate and post a comment

(76)
(14)

Hector Vargas

re: OEC 9400 Flipper Stuck

July 04, 2017 12:47

You have to open the Colimator, there is a solenoid inside with a micro switch that senses the position of the flipper, some times the solenoid is stuck, so you have to clean it, some times the switch is damaged and you have to replace it.

Turn on the C-arm, you have to listen a sound, like clack, that means that the flipper has movement, you can see that movement if you place the colimator off to see inside where the flipper is.

hectorx@live.com.mx

Log inor Register

to rate and post a comment