Optimus 50 bucky does not work

Tags: Radiation Therapy, PHILIPS

(2)

MOUADH Alshabi

Optimus 50 bucky does not work

January 23, 2023 10:29

I face problem in optimus 50 bucy does not work
when power up unit
All LEDs on control panel are glow up and at the generator side all boards "Red lEDS" are glow up. except the led on the 102 Universal I/O Interface it still ON and all leds are turn off except the led on 102 Universal I/O Interface.
When i expose with free cassette every thing is ok but when i try to use the bucky it be ready but there is no exposer
I need Service manual too
Thanks

Log inor Register

to rate and post a comment

(172)
(25)

Tibor Ivanszky

re: Optimus 50 bucky does not work

January 24, 2023 10:27

Every time you start the system you have to activate the bucky by pulling the tray out and pushing back.

Log inor Register

to rate and post a comment

(2)

MOUADH Alshabi

re: Optimus 50 bucky does not work

January 24, 2023 11:14

Prepration circule also appear on the control panal but there is no exposure

Log inor Register

to rate and post a comment

ISS - Contact Us Today For All Your Contrast Injector Needs: 888-667-1062

Your comprehensive contrast injector solution for 10 years to include: on-site service, telephone technical support, depot repair with loaner options, parts sales, injector sales, injector install & de-install, Level 2 & beyond training.

advertisement

Upcoming Auction Events

Tekyard Medical
TIMED Ending Soon
Feb 08, 2023 (219 lots)
Global Medical Devices LLC
TIMED
Feb 10, 2023 (83 lots)
SDI Medical Consultants
TIMED
Feb 13, 2023 (116 lots)
Unitedrecovery.ca
TIMED
Feb 14, 2023 (72 lots)
Therapeutic Technologies
TIMED
Feb 17, 2023 (2053 lots)