PHILIPS BRILLIANCE 16 Air Critical Error

Tags: CT, PHILIPS

(13)
(2)
(1)

Ricardo Carvalho

PHILIPS BRILLIANCE 16 Air Critical Error

November 22, 2021 11:16

I'm getting an intermittent error mostly when couch up/down movements are performed, but sometimes it occurs without movements, even when nobody is using the equipment. When error occurs the Gantry turns off and I get the critical error:

code: 33227254
Message: S_LIFESYNCMGRTASK_LIFESYNC_FAIL

This error is very intermittent, sometimes occurs several times a day, and sometimes the equipment is working many weeks without troubles.

Someone can give me any kind of help.

Log inor Register

to rate and post a comment

(6)

rockbob

re: PHILIPS BRILLIANCE 16 Air Critical Error

November 24, 2021 12:40

Check safety switch and cabling for the table brace. cable bends every time table is raised or lowered.