Hi!
I have a problem with Philips Envisor: black line on the image on all probes.
I have bad CMOS battery and changed it, after i changed date and time in bios.
But it not helped.
I do all tests and its ok.
Any ideas?
Thanks

Log inor Register

to rate and post a comment

(53)
(4)
(29)

Brian Seltzer

re: Philips Envisor black line on the image

February 14, 2018 09:28

Looks like you might have a tr board bad.
We can replace if needed

Log inor Register

to rate and post a comment

(73)
(4)

ultrasoundman

re: Philips Envisor black line on the image

February 14, 2018 09:28

Need to see a picture, sounds like receiver

Log inor Register

to rate and post a comment

(113)
(54)
(5)

Jim Carr

re: Philips Envisor black line on the image

February 14, 2018 09:29

Does it appear to be dropout due to dead transmitters or receivers? Did you try swapping the positions of the two TR boards?

Log inor Register

to rate and post a comment

(1)

Sener MUTLU

re: Philips Envisor black line on the image

February 14, 2018 09:30

U hve any picture?

Log inor Register

to rate and post a comment

(55)
(5)

Ramil Mirshanov

re: Philips Envisor black line on the image

February 14, 2018 09:30

Hello! Aleksander, it is necessary to clarify that the black line goes only in the scan sector ?

Log inor Register

to rate and post a comment

(144)
(21)

Aleksander Proof

re: re: Philips Envisor black line on the image

February 14, 2018 02:46

Black line only scan sector. And it repeats the shape of the beam curved or straight
I swap tr board and image is the same
In test all boards are ok

Log inor Register

to rate and post a comment

(145)
(31)

Kobris

re: Philips Envisor black line on the image

February 14, 2018 09:30

please send photo of the screen, all probes. kobris@gmail.com

Log inor Register

to rate and post a comment

(144)
(21)

Aleksander Proof

re: re: Philips Envisor black line on the image

February 15, 2018 09:27

Hi!
Problem was really in TGC, its low settings. I moved these sliders every few times and the image recovered.
Thank you all for your help!

Log inor Register

to rate and post a comment

Your #1 Source For Everything Ultrasound!

Redstone Healthcare - Exceptionalism is what we do. Offering new and reconditioned ultrasound systems, transducers, parts and service nationwide from machines to transducers, C-arms and more! CALL (773) 294-5030 DOTmed Certified DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED