Philips optimus 65 not giving exposures

November 22, 2021 11:28

I am looking for help in fixing a Philips optimus 65 xray system.
The green LED ( circular) is not turning on on the control pane. Kindly some with vast knowledge help out.

Log inor Register

to rate and post a comment

(269)
(40)

CRAIG WAMBURA

re: Philips optimus 65 not giving exposures

November 23, 2021 05:02

Does it light up yellow or red?
Does the anode rotate?
What error, if any, do you get?

Could be tube, HT cable or converter
I can help

Log inor Register

to rate and post a comment

(3)
(1)

Morris Chimuti

re: re: Philips optimus 65 not giving exposures

November 24, 2021 12:40

All lights are behaving fine, the only light not powering is the ready light

Log inor Register

to rate and post a comment

(40)
(8)

RobertCow

re: Philips optimus 65 not giving exposures

November 24, 2021 10:39

please tell the exposure condition, table, bucky, free exposure?
press the Test key on the CS, a message shows with the problem of the not ready light.

Log inor Register

to rate and post a comment

(70)
(11)

fortronik

re: Philips optimus 65 not giving exposures

November 24, 2021 10:41

You have not met the conditions for taking the photo
- anti-scatter grill not ready
- if you have a CAN system, there may be a problem with the actuators
- if you have a CAN system then there is a TEST button on the lamp holder, will prompt what is not fulfilled until the camera is ready

Log inor Register

to rate and post a comment

Auction Alert - Live Now: ZRG Medical Surplus Auctions December 3rd

ZRG Medical closeout auctions take place on DOTmed.com. Bid today on 122 lots of medical equipment. Onsite or virtual inspection available. Please contact 760-438-8835 ext. 100 for an appointment.

DOTmed Certified DOTmed 100
advertisement

 

RSTI Newsletter

Upcoming Auction Events

KEEBOMED
TIMED
Nov 30, 2021 (1646 lots)
Tekyard Medical
TIMED
Dec 01, 2021 (361 lots)
ZRG Medical
TIMED
Dec 03, 2021 (122 lots)
Primis Medical
TIMED
Dec 09, 2021 (564 lots)