> Please help with Agfa CR35 X

Please help with Agfa CR35 X

Tags: PACS / DR / CR / Digital

karim abdelkarim

Please help with Agfa CR35 X

March 12, 2018 09:48

We have a problem with the agfa CR 35 X . it gives the error code 20442 , and the image looks distorted with a lot of horizontal lines .
This Error occurs occasionally .

Log inor Register

to rate and post a comment

Zhang r.s

re: Please help with Agfa CR35 X

March 13, 2018 09:06

(1) Enter the Service menu

(2) Select
4 Show error
1 Explain error code

(3) Select in error code history corresponding entry
(4) Reason, meaning and cure of error are listed

Log inor Register

to rate and post a comment

(314)
(58)
(2)

juan pablo amado

re: Please help with Agfa CR35 X

March 13, 2018 09:07

20442
E_SCD_SCB_DMA_BUF_OVERFLOW
Meaning : There was at least one DMA ring buffer overflow during scan.
Reason : The image data was not stored away quick enough by the
following buffer acquisition (BAC) software.
Bad disk, system overload, ...

Log inor Register

to rate and post a comment

(13)

hasan raza

re: Please help with Agfa CR35 X

May 30, 2018 09:40

contact me my email address faisal.medical.equipments@gmail.com

Log inor Register

to rate and post a comment

OiHealthcare - ISO 13485 Certified - Call 1-888-278-9933

Oxford Instruments Healthcare, is a leading ISO 13485-certified organization, that specializes in providing quality refurbished CT & MRI systems, service, mobile units & parts at prices you can afford. Call 888-278-9933 Email DOTmedinfo@oxinst.com

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED