Polydoros IT

January 09, 2018 02:12

I at all!! I have some problem about a Polydoros IT generator. Often appair error 32 and 001 on consolle. On service manual say’s replace D500,, have you got some ideas about? Thanks in advance!!!

Log inor Register

to rate and post a comment

(12)
(2)

stefano bandini

re: Polydoros IT

January 10, 2018 09:17

Hi Vincenzo!
Contact me on mail, i think that i can help you.

Log inor Register

to rate and post a comment

(47)
(9)

fortronik

re: Polydoros IT

January 10, 2018 09:18

Check the 5.1V power supply. Very often these power supplies are damaged.
Check if the power supply is on the D500 is 5.1V
You may have evil contact in the fuse socket for 5.1V there is a big current

1 person liked this
Log inor Register

to rate and post a comment

(4)

Vincenzo Conte

re: re: Polydoros IT

January 12, 2018 09:38

Very interesting !! Thanks for your kind answer !