Problem in GE VCT/ 750HD CT scanner

November 18, 2022 10:19

Dear All,
We have problem in GE CT 750HD, error number 260107251, (Axial Motor Control AMC logic lost communication with the Axial Drive ( Communication watchdog failure)
The system is reporting this error upon start up, while the hardware reset successfully with no errors.

When we asked for tube warm up the gantry (AXIAL) rotated normally but with no x ray as the system is already reported the above error previously.

Can you please help us in investigation?

Log inor Register

to rate and post a comment

(6)

AMC

re: Problem in GE VCT/ 750HD CT scanner

November 21, 2022 10:38

Try replacing Axial encoder

Log inor Register

to rate and post a comment

(6)

G2DIAZ

re: Problem in GE VCT/ 750HD CT scanner

November 21, 2022 10:39

Hi! Do you check the axial encoder and axial pulse to see is OK?

Log inor Register

to rate and post a comment