Problem with a GE Advantx R/F with a MPPU Generator

January 08, 2020 10:12

Dear Colleagues:
I am having a problem with a GE R/F Advantx with a MPPU Generator.
Error 42C0 comes up on the control console saying this is a communication problem.
The error appears just after turning on the machine and it goes away by itself after the machine is on for a little while.
I really need someone that knows profoundly this machine and could give me a good insight of what could be the cause of this problem.

Sincerely:

Manuel (Manny) Brugat/Image Instrumentation, Inc.

1 person liked this
Log inor Register

to rate and post a comment

(5)
(1)

MRISAGE

re: Problem with a GE Advantx R/F with a MPPU Generator

January 09, 2020 09:09

Does the system operate normally?

Log inor Register

to rate and post a comment

(102)
(9)
(6)

Manuel Brugat

re: re: Problem with a GE Advantx R/F with a MPPU Generator

January 15, 2020 10:28

MRISAGE:

The system after several or a lot of cycling, finally it works normally, but it takes a long while.
Do you have an explanation for this?

Log inor Register

to rate and post a comment

(5)
(1)

MRISAGE

re: re: re: Problem with a GE Advantx R/F with a MPPU Generator

January 21, 2020 10:26

It could be a power supply in console. The best bet is to connect to the Advantx with a laptop and get the error log. Does the error code list a Node?