Problem with CT Siemens Somatom Spirit

February 08, 2019 09:08

hi, everyone.
I have problem with Siemens Somatom Spirit. Sometimes when start exam make topogram and on 3-4 axial scan system stop scan without any error on display . Wait 10-15 second and start again. On eventlog I found error CT_GPC_E_10_42_P1P2
Thanks

1 person liked this
Log inor Register

to rate and post a comment

(27)
(4)

MIGUEL SAYRITUPAC

re: Problem with CT Siemens Somatom Spirit

February 10, 2019 01:30

Hello George.
Write to me miguel.araujo2270@gmail.com.
To atach pic of event log.

Best regards.
Miguel Angel

1 person liked this
Log inor Register

to rate and post a comment

(274)
(48)

Gabriel Cool

re: Problem with CT Siemens Somatom Spirit

February 10, 2019 01:31

Could you send me photo of error on email?

1 person liked this
Log inor Register

to rate and post a comment

(12)
(1)

Hesham Rihan

re: Problem with CT Siemens Somatom Spirit

February 10, 2019 01:33

can you send me on hesham.rihan@imagemedical.net a snapshot of the event log or even an extract of it and i will try to help

1 person liked this
Log inor Register

to rate and post a comment

(17)

Juan Jose Garate Pirela

re: Problem with CT Siemens Somatom Spirit

February 15, 2019 09:11

GPC means that you have a problem with Gantry Panel Control, it may opens the stop loop. Also sometimes this problems happens when the Sprocket coupled to the gantry that counts its rotation has some Teeths missed.

Log inor Register

to rate and post a comment

(7)
(1)

Georgi Gaydov

re: re: Problem with CT Siemens Somatom Spirit

March 12, 2019 01:19

Hi!

Sorry for may late reply.

The problem is on anode rotation of x ray tube.

The tube have very big noise and vibration when turn on with 60HZ/150HZ and sometimes the tilt sensor is changing to +1;0;-1 and stop scan.

Log inor Register

to rate and post a comment

Marquis Medical - Experts in SIEMENS & GE PET & PET/CT Service & Parts

Our engineers are trained strictly on Siemens & GE PET to Guarantee an expert on all service calls. Our large inventory of PET & PET/CT systems & parts allows us to provide a TRUE Turn-key solution. Ph:866-645-7457 DOTmed Certified

 

advertisement

 

 

advertisement

 

advertisement
advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
ALL RIGHTS RESERVED