SEDECAL SHF 330 communication error E74

Tags: Radiology - R/F

(21)
(3)

Emmanuel Quaye

SEDECAL SHF 330 communication error E74

September 11, 2020 09:38

SEDECAL SHF 330. I have communication error E74 After exiting service Mode.

1 person liked this
Log inor Register

to rate and post a comment

(22)
(1)
(1)

Howard Karp

re: SEDECAL SHF 330 communication error E74

September 14, 2020 10:31

I looked up Error 74: excessive current in main winding braking at 3300 rpm-CHECK CR8, 9, 10, 11 in Interface DRAC-HF. Also check Rotor wiring for shorts to ground.