Service Manual And Helpe Esu Itc Model Itc 100-D - Service Technicians Forum
advertisement
Current Location:
>
>
> service manual & helpe ESU itc model itc 100-d

service manual & helpe ESU itc model itc 100-d

(8)
(5)

ahmad badi

service manual & helpe ESU itc model itc 100-d

August 13, 2019 03:19

we have ESU ( itc co model itc100d )
its powered on but not work for all modes cutting or coagulation or bipolar ( no out but )
i open the unit and i find resistor no : R12 burned as As the attached picture
so if you can help me : as to sending me Resistance value R12 value
or send the schematic or service manual>>>> pleas help me

1 person liked this
Log inor Register

to rate and post a comment

(8)
(5)

ahmad badi

re: service manual & helpe ESU itc model itc 100-d

August 14, 2019 10:11

R 12 in the PSB

1 person liked this
Log inor Register

to rate and post a comment

(8)
(5)

ahmad badi

re: service manual & helpe ESU itc model itc 100-d

August 15, 2019 10:05

pleas help:
any one have a service manual for Elector surgical unit ITC -100D or schematic

1 person liked this
Log inor Register

to rate and post a comment

(8)
(5)

ahmad badi

re: re: service manual & helpe ESU itc model itc 100-d

August 18, 2019 11:28

its finish the problem
R12 =70 ohm
now my ESU is OK

Log inor Register

to rate and post a comment

(1)
(1)

syr70

re: service manual & helpe ESU itc model itc 100-d

August 15, 2019 10:05

re: re: service manual & helpe ESU itc model itc 100-d
i need it too !!!!!!!!

1 person liked this
Log inor Register

to rate and post a comment

All Original Siemens & Philips Parts & Tubes Available at Attractive Prices

All parts available upon request. Please include part number and description for your request. Warranty available. Call 813-925-1314 or email at Sales@Elsmed.com DOTmed Certified DOTmed 100

 

advertisement

 

advertisement

 

advertisement

 

advertisement
advertisement