Hi everybody !
How could I find a Service Manual for Sysmex K-1000 and a User Manual for Coulter MAXM ?
Thanks in advance !

Log inor Register

to rate and post a comment

(7)
(3)

Medimarc

re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 01, 2014 08:39

I can give you the manual for the Coulter per email.

2 people liked this
Log inor Register

to rate and post a comment

(5)

Christos Paschalis

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 01, 2014 04:32

Dear Medimarc, thank you very much ! My email is : chris7pas@gmail.com.

Log inor Register

to rate and post a comment

(14)
(3)
(12)

JMScientific

re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 01, 2014 08:39

Hello, I have both of the manuals you are looking for. I will give you both of them burnt on a cd for $150.

Log inor Register

to rate and post a comment

(49)
(8)

Ricardo Rivera

re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 01, 2014 06:54

WHAT PROBLEM YOU HAVE IN K1000

Log inor Register

to rate and post a comment

(5)

Christos Paschalis

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 02, 2014 09:14

I must change the tubings (very old). So I will need tubing diagrams, in case of mistake.

Log inor Register

to rate and post a comment

(14)
(3)
(12)

JMScientific

re: re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 02, 2014 10:03

Hello, I also have all the tubing, filters, and diagrams for both units. I will send the diagrams and manuals with the purchase of the tubing and filters at no extra charge.

Log inor Register

to rate and post a comment

(49)
(8)

Ricardo Rivera

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

July 02, 2014 04:35

I recommend that change one by one

Log inor Register

to rate and post a comment

(4)

Ehsan Mosleh

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

February 12, 2019 10:07

Hi i try to service sysmex k800 it gives no diluent replenish & press1 ;i check srv and valve and diaphragm also diluent is there, l check for leakage, now i think tube connection is not correct, can u help me with this and if u hsve tube diagram send it to me. Thanking you

Log inor Register

to rate and post a comment

(4)

Ehsan Mosleh

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

February 12, 2019 10:07

Hello

Log inor Register

to rate and post a comment

(4)

Ehsan Mosleh

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

February 13, 2019 09:53

Hi i try to service sysmex k800 it gives no diluent replenish & press1 ;i check srv and valve and diaphragm also diluent is there, l check for leakage, now i think tube connection is not correct, can u help me with this and if u hsve tube diagram send it to me. Thanking you

Log inor Register

to rate and post a comment

(4)

Ehsan Mosleh

re: re: Service Manual Sysmex K-1000 , User Manual Coulter MAXM

February 13, 2019 11:52

Hi i try to service sysmex k800 it gives no diluent replenish & press1 ;i check srv and valve and diaphragm also diluent is there, l check for leakage, now i think tube connection is not correct, can u help me with this and if u hsve tube diagram send it to me. Thanking you

Log inor Register

to rate and post a comment

Brandon Transfer & Storage - Rigging, Moving & Set Up For Medical Equipment

Family owned since 1919, for your next medical move, rely on the specialists at Brandon Transfer & Storage. For references from the medical community & a free moving estimate, call us at 561-683-5000 and we care! Visit www.brandontransfer.com Today!

 

advertisement

 

advertisement

 

advertisement
Advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
ALL RIGHTS RESERVED