Service software for Siemens Siremobil Compact L

July 19, 2018 08:15

Hello all! I have two error messages in a Siemens Siremobil Compact L. The first error appears during the system startup, and is the error 7400, shown on the panel's display (not shown on the monitor). If you press any key the error is cleared and will be possible to perform exposures. After each exposure, the error 7409 appears on the panel's display. You must press any key to clear the error and make new exposures. I found in the service manual informations related to these errors. The erros may be related to iris pot calibration. But it is needed a service software with password to perform these calibrations and other functions. This software is to be instaled on a service pc and connected to the Siemens Siremobil. Can somebody help me with this service software, password and some explanation about the use, or about the error? My email is sanmedicalbr@gmail.com. Thankyou very much!

Log inor Register

to rate and post a comment

(30)
(9)
(28)
(50)

Bill Masters

re: Service software for Siemens Siremobil Compact L

July 19, 2018 05:41

Error 7409

Meaning:
TV-Iris wrong position

Possible problem:
If the error 7400 to 7408 were detected, this error is displayed after each radiation release.

User warning:
The user must delete this error before radiation can be released.

Possible help:
For more information, read out the previous error messages.
Check the doserate adjustment.
Check the actual value potentiometer.
Check the actual value potentiometer wires.

Log inor Register

to rate and post a comment

(260)
(38)

CRAIG WAMBURA

re: re: Service software for Siemens Siremobil Compact L

May 04, 2021 11:27

I have a siremobil with error 5014 and 5015 at startup

Log inor Register

to rate and post a comment

(39)
(1)

LUIS ERNESTO OROZCO

re: Service software for Siemens Siremobil Compact L

July 21, 2018 01:16

Your problem is maybe your Cam-Iris motor, we recently repair a system with the same error, if you want more help email me servicio@saisco.com.mx

Log inor Register

to rate and post a comment

(341)
(79)
(13)

Manuel Casanova

re: Service software for Siemens Siremobil Compact L

July 23, 2018 05:58

Error E7409 TV-Iris wrong position

Possible trouble: If the error 7400 to 7408 were detected, this error is
displayed after each radiation release.
User warning:
The user must delete this error
before radiation can be released.

Possible help: For more informations, read out the previous error messages.
Check the doserate adjustment.
Check the actual value potentiometer.
Check the actual value potentiometer wires.

Log inor Register

to rate and post a comment

Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students.

DOTmed Certified
advertisement

 

Hitachi

 

RSTI Newsletter

Upcoming Auction Events

Therapeutic Technologies
TIMED Ending Soon
May 07, 2021 (177 lots)
Premier Medical Depot
TIMED
May 12, 2021 (595 lots)
ZRG LLC.
TIMED
May 14, 2021 (162 lots)
Surgamed
TIMED
May 24, 2021 (329 lots)