Hi to all,

Anyone who can help me in looking the meaning of the Error of the Siemens Siremobil Compact L as follows;

1. E2008
2. E5019
3. E5020

Thank you!

Log inor Register

to rate and post a comment

(251)
(37)
(28)

Wayne Horsman

re: Siemen Siremobil Compact L E2008, E5019, E5020

August 11, 2019 02:24

Error W2008 = Service, ser. receive line error
Possible trouble = The receive line has a constant level.
Possible help = Check receive line between D1 and service interface. Possibly service-PC, serial interface or D1 defective.
---------------
Error E5019 = Gen. kV/mA courve selection, int. error
Possible trouble = Generator registered the selection of an inadmissible KV/mA curve.
Possible help = Check parameters for KV/ma curves. D1 defective.

Log inor Register

to rate and post a comment

(338)
(75)
(13)

Manuel Casanova

re: Siemen Siremobil Compact L E2008, E5019, E5020

August 11, 2019 02:25

Salvador:

W2008: Possible trouble: The receive line has a constant level.

Possible help: Check receive line between D1 and service interface.
Possibly service-PC, serial interface or D1 defective.

E5019: Possible trouble: Generator registered the selection of an in admissible
KV/mA curve.

Possible help: Check parameters for KV/ma curves.
D1 defective.
E5020: is not found.