dismiss

Clean Sweep Live Auction on Thur. March 28th. Click to view the full inventory

SEARCH
Current Location:
>
>
> Siemens Acuson X300 no boot

Siemens Acuson X300 no boot

(4)
(1)

Masoodtech

Siemens Acuson X300 no boot

July 03, 2018 08:22

I have problem with Siemens X300. When I press the power on button the machine tries to boot but within 2-3 seconds shuts down and restarts the power-up sequence again, it keeps repeating until I forcefully shut it down.
The AC power supply only shows power light and battery light solid green, none of the other lights come on, 48V and 12V checked out OK on AC.
Also checked the DC-DC TPs:
All tested OK but TP11 and TP12 polarities are reversed,
reading:
TP11 +14.2
TP12 -14.2
I used the test reference from another posts by Anthony Ngo(Many thanks) but is this a typo or am I missing something.
Also can somebody please email me the service manual to masoodexch@gmail.com.

Sorry for the long post.




1 person liked this
Log inor Register

to rate and post a comment

(17)
(4)

Abhishek Jain

re: Siemens Acuson X300 no boot

July 07, 2018 10:54

TP11 and TP12 voltage you mentioned are correct.
Follow these steps to troubleshoot and see if the symptom changes
1. Remove all transducers and boot up machine. Check if symptom changes.
2. Remove TR board and boot up machine. If the symptom is same refit it.
3. Remove RC board and boot up machine. If the symptom is same refit it.
4. Remove Hard disk and DVD drive connection. See if symptom changes. If it doesn't then refit them.
5. Remove TI module and boot up machine. See if symptom changes. If it doesn't then refit them.

Let me know after you perform all above steps. Then I will tell you whats the problem is.

Log inor Register

to rate and post a comment

(4)
(1)

Masoodtech

re: re: Siemens Acuson X300 no boot

July 08, 2018 12:10

Tanks for taking time, Will check and let you know

Log inor Register

to rate and post a comment

(4)
(1)

Masoodtech

re: re: Siemens Acuson X300 no boot

July 08, 2018 12:11

Abhishek, checked everything you suggested in the sequence. no change at all.
I also checked and verified the TPs on the BE board, all good, also checked cooling fans, both work fine.
Any help is appreciated.

Log inor Register

to rate and post a comment

(17)
(4)

Abhishek Jain

re: re: re: Siemens Acuson X300 no boot

July 11, 2018 10:38

If the steps I mentioned did not change the symptom then its most probably a problem of RM (ETX or CPU) board which is attached to BE board. Let me know if that solves problem. Message me on abhi21687@gmail.com

Log inor Register

to rate and post a comment

Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students.

 

 

advertisement

 

advertisement

 

advertisement
Advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
ALL RIGHTS RESERVED