Hi, everybody
I have a Siemens Acuson X300 that no turn ON. Display on monitor "Check Signal Cable".
Troubleshooting show the I/O output is bad. But the troubleshooting show that the RM Board is bad too.
Somebody have this proble with this Ultrasound X300

Thanks in advance

Regards

Log inor Register

to rate and post a comment

(118)
(21)

simo elbrami

re: Siemens Acuson X300

April 17, 2019 08:57

Hello juan, its a powersupply problem

Log inor Register

to rate and post a comment

(88)
(15)

Juancho

re: re: Siemens Acuson X300

April 18, 2019 12:27

Hi, Simo I have other X300 but this show "the system is shutting down automatically to prevent permanent damage.please contact Siemens Service" message. What do you think about it?
regards

Log inor Register

to rate and post a comment

(2)

Kekchiew Bong

re: re: re: Siemens Acuson X300

September 12, 2019 09:09

1st: check is there any orange indicator on power supply underneath the ultrasound. 2nd: check is the ventilation fans working and filter is clean?

Log inor Register

to rate and post a comment

(2)

Kekchiew Bong

re: Siemens Acuson X300

September 12, 2019 09:09

Hi, have you manage to solve the problem? I had X300 display on monitor show: Check Signal Cable but when i connect monitor to VGA, there is images.

Log inor Register

to rate and post a comment

(88)
(15)

Juancho

re: re: Siemens Acuson X300

September 13, 2019 10:43

Hi, Kekchiew
you must connect the backup monitor directly to the video cable of the main monitor and check.
let me know the test results.
Regards

Log inor Register

to rate and post a comment