Good afternoon; I need to know how to start the Siemens Acuson X300 power supply without connecting it to the equipment.
Buenas tardes; necesito saber cómo arrancar la fuente de poder de Siemens Acuson X300 sin conectarla al equipo.
Gracias

Log inor Register

to rate and post a comment

(6)
(1)

Hector Araujo

re: Siemens Acuson X300

July 20, 2021 10:50

I solved it. I repaired the AC-DC power supply.
Thanks

Log inor Register

to rate and post a comment

(2)

Sam Narcizo

re: re: Siemens Acuson X300

April 12, 2022 10:34

Hello Hector!
Can you explain us how you did It? Im also trying to start It without the Full equipment

Log inor Register

to rate and post a comment

(2)

Sam Narcizo

re: re: Siemens Acuson X300

April 12, 2022 10:34

hello Hector!
can you explain us how you did it? i'm also trying to simulate the machine on power supply

Log inor Register

to rate and post a comment

(12)
(1)

Jon Himavat

re:re: re: Siemens Acuson X300

May 05, 2022 09:17

normally tgere will be a triger in power supply
you just short ground black lead with that trigger to start.
normally this happens when u press power button motherboard triggers it. but it can be done manually if you can identify it.

Log inor Register

to rate and post a comment

(29)
(6)

MTC

re: re:re: re: Siemens Acuson X300

January 23, 2023 10:32

Dear Jon,

I have same issue on my X700.
Can you help ?

Log inor Register

to rate and post a comment

(23)
(3)

UltraServiceMed

re: re: re:re: re: Siemens Acuson X300

January 24, 2023 05:51

u need to find the trigger normally it can be identify if you have power supply data sheet or if u can trace the connectors. Normally computer repair centers can help with this .
u need to short trigger to a ground to start it.
Normally this action is done by motherboard. till that time power supply will only give supply to mst

Log inor Register

to rate and post a comment

Auction Alert - Live Now: Tekyard Medical Surplus Auctions February 8th

Tekyard Medical closeout auctions take place on DOTmed.com. Bid today on 221 lots of medical equipment. Onsite or virtual inspection available. Please email at auctions@tekyard.com for an appointment.

DOTmed Certified DOTmed 100
advertisement

Upcoming Auction Events

Tekyard Medical
TIMED Ending Soon
Feb 08, 2023 (220 lots)
Global Medical Devices LLC
TIMED
Feb 10, 2023 (83 lots)
SDI Medical Consultants
TIMED
Feb 13, 2023 (116 lots)
Unitedrecovery.ca
TIMED
Feb 14, 2023 (72 lots)
Therapeutic Technologies
TIMED
Feb 17, 2023 (2053 lots)