Need service program manual of siemena mammomat nova 3000.

I have error code 802 to resolve.

2 people liked this
Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: Siemens mammomat nova3000

October 29, 2019 10:37

Try swapping the Bucky's around . If the problem goes away, the problem could be the bucky switch or motor. The switch was a common problem back in the day. Siemens even came out with a kit to repair it.

Log inor Register

to rate and post a comment

(103)
(21)

Germni

re: Siemens mammomat nova3000

October 29, 2019 10:47

Hello,
Code Error 802 OKT 2 HAS RESPONDED
Presure switch on tube housing or beam form anode or door switch open
I mind your tube is active pressure switch

Log inor Register

to rate and post a comment

(67)
(9)

Ramil Mirshanov

re: Siemens mammomat nova3000

October 29, 2019 10:47

Hi,
Er 802
OKT 2 has responded
Displayed on PC - OKT pressure switch on tube housing or beam form anode or door switch.
PH1 is replaced with OKT2, HV tank with tube housing.
BestRegards.

Log inor Register

to rate and post a comment

(56)
(18)

Mikhail Zakurdaew

re: Siemens mammomat nova3000

October 29, 2019 10:48

Er 802 “OKT 2 has responded”
Displayed on PC: OKT2 pressure switch on tube housing or beam form anode or door switch
Description Same as Er 630 (PH1 is replaced with OKT2, HV tank with tube housing).
Er 630 “Pressure switch PH1 has responded”
Displayed on PC: PH1 pressure switch on the HV tank activated
Description PH1 pressure switch on the HV tank is activated, the above-mentioned error is displayed.
Fault elimination Overload of the tube assembly or of Hl.
Line interruption (see wiring diagram).
Board D702 defective (monitoring).

Log inor Register

to rate and post a comment

(5)

Mauro Stempelatto

re: Siemens mammomat nova3000

November 01, 2019 11:27

Error 802 "OKT 2 has not responded"
"Okt2 pressure switch on tube housing or beam form anode or door switch"
Same as Er 630 (Replace PH1 with OKT2, hv tank with tube housing)

Fault Elimination:
Overload of the tube assembly or of HI
Line interruption (see wiring diagram)
Board D702 defective (monitoring)

Log inor Register

to rate and post a comment

(2)
(2)

Rahim Rajab Ali

re: re: Siemens mammomat nova3000

November 08, 2019 10:38

For error 802, all check has done, now I am going to check board d702. All other things were fine but now I am facing error 302 and 620.

Now what should I do?

Log inor Register

to rate and post a comment

(103)
(21)

Germni

re: re: re: Siemens mammomat nova3000

November 10, 2019 11:36

Er 620
“UCTRL is missing”
Displayed on PC: U
Description
Master D702 monitors the control voltage of the transistor module in the inverter (+ 15V).
If this voltage fails, the above-mentioned error is displayed.
Fault elimination
Fuse F5 defective (see wiring diagram).
Measure voltage on T2.
Board D702 or D710 defective.
Er 630

anst
+15V to control inverter not present
Er 302
“Filament current monitoring I
Displayed on PC: I
Description
Hmin
has responded.”
During “Stand-by”, the master checks whether there is any filament current. If the filament
current does not reach a predetermined value, the above-mentioned error is displayed.
Fault elimination
Hmin
- filament current too low at standby
Check fuse F3 on T2 (see wiring diagram).
Check X8 connector (both on board D700 and on the cable).
Check filament wires.
Filament of tube assembly defective.
PC board D705 defective.
your tube is with trouble

Log inor Register

to rate and post a comment

(1)

Ali28

re: re: re: re: Siemens mammomat nova3000

December 09, 2019 10:41

Thanks, problem has resolved but when do exposure error 801 occurred

All is well but exposure is not working.