Siemens Somatom Sensation 64 problem

Tags: CT, SIEMENS

(161)
(37)
(3)

Dmitry Dederer

Siemens Somatom Sensation 64 problem

April 28, 2014 08:53

Hello! I have the problem with siemens Somatom Sensation 64. CT stops the scan in CarotidAngion research with bolus. Another protocols works without the errors and failures. There are the following errors in EventLog in the moment of the scans stopped.
3107 CT_MCU Control error MCU (E 00 01 6b 01 2b 00 00)
3106 CT_MCU Control error MCU (E 00 01 6b 01 2b 00 00)
41 CT_MSM Abort by control system (AD): Detail see abort reason CT_MCU 3106
301 CT_ITH Scan abort due too many defective readings (HV drop)
1 CT_ITH IRS error information: AQP: number of defective readings out of range because of tube arcs 0x00800000
3042 CT_XRS Control error XRS (E 50 01 2a 00 00 00 00)
What is the problem? Tube, HV-tank, detectors or the error in the protocol?

Log inor Register

to rate and post a comment

(2681)
(619)
(17)

Lenny Place

re: Siemens Somatom Sensation 64 problem

April 29, 2014 08:42

Dmitry
Have you tried to run the tune up GENOSC, then the gettering? it is very possible that is is a tube arc (how old is the tube and how scan seconds?). have you checked the brushes and slip rings. Gettering sometimes will condition the tube (sometimes key word)

Log inor Register

to rate and post a comment

(161)
(37)
(3)

Dmitry Dederer

re: re: Siemens Somatom Sensation 64 problem

April 30, 2014 08:37

Hello, Lenny! I didn't run tune up. The CT is 2008 year of manufacturing. We cleaned the slipring ang check the brushes. What can we do else?

Log inor Register

to rate and post a comment

(2681)
(619)
(17)

Lenny Place

re: re: re: Siemens Somatom Sensation 64 problem

April 30, 2014 01:29

I would run the GENOSC and gettering

Log inor Register

to rate and post a comment

(57)
(3)

Mr.CT.service

re: Siemens Somatom Sensation 64 problem

April 29, 2014 08:44

Dmetry,

If System does not arc at all with other protocols even with secuencial scans or warm up do a system backup. Then do restore of protocols only.
Is it a reinstallation?
Have you tried gettering?

Good Luck

Log inor Register

to rate and post a comment

(161)
(37)
(3)

Dmitry Dederer

re: re: Siemens Somatom Sensation 64 problem

April 30, 2014 08:37

Hello, Mr. CT service! Other protocols works well, except CarotidAngio with contrast injection. We reinstall the software from cold and restored the backup but the priblem still persist. What can we do else?

Log inor Register

to rate and post a comment

(57)
(3)

Mr.CT.service

re: re: re: Siemens Somatom Sensation 64 problem

May 01, 2014 08:47

Exactly as Lenny.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: Siemens Somatom Sensation 64 problem

April 29, 2014 08:44

Hi Sir,
It looks TUBE problem. you need to do some KV tests to confirm it. Do you have level 7 password?
Do you have KV tester tool(KV pluger)?
Regards
Lin Dong

Log inor Register

to rate and post a comment

(161)
(37)
(3)

Dmitry Dederer

re: re: Siemens Somatom Sensation 64 problem

April 30, 2014 08:37

Hello, Lin! We have passwords and can have the HV plugs for simulate the filament.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: re: re: Siemens Somatom Sensation 64 problem

April 30, 2014 01:32

email me: lindong123@hotmail.com

Log inor Register

to rate and post a comment

(2)

Max Saran

re: Siemens Somatom Sensation 64 problem

July 17, 2015 02:58

I have a XRS 3049 error on a Siemens Sensation 64. GENOSC worked but Gettering reports this XRS 3049 error. Any clue?

Log inor Register

to rate and post a comment

(2681)
(619)
(17)

Lenny Place

re: Siemens Somatom Sensation 64 problem

July 17, 2015 03:23

Possibly the tube

Log inor Register

to rate and post a comment

(2)

Max Saran

re: re: Siemens Somatom Sensation 64 problem

July 17, 2015 06:14

Yes it might be the tube. but any idea what that means?

Log inor Register

to rate and post a comment

(57)
(3)

Mr.CT.service

re: re: re: Siemens Somatom Sensation 64 problem

July 18, 2015 11:48

Results of HVC0x tests?
have you just replace e-box?
Try with new HT cables?
Any noice (arc) while GENOSC?

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: re: Siemens Somatom Sensation 64 problem

July 17, 2015 06:14

how many scan*s is your tube ?
You have better use Level 7 passowrd to open the Error log, and see the cause reasons, test steps,results....

Log inor Register

to rate and post a comment

Benedito de Vasconcelos Gomes JĂșnior

re: Siemens Somatom Sensation 64 problem

January 13, 2021 09:52

Hello! Did anyone found a solution? I have the same problem. But on teste, the is no arcs on tube, it has 120.000 scans. Thanks!