Siemens Sonoline G40 - Service Technicians Forum
advertisement

Siemens Sonoline G40

Tags: Ultrasound, SIEMENS

(8)

gavin Sprong

Siemens Sonoline G40

May 22, 2019 09:57

I have a fault on a iemens G40 where on Self Diagnostics it brings up CHECK SIGNAL CABLE - NO SIGNAL.Aetr anout a minute the unit then just switches itself off. Does anyone have a service manual for this unit you could mail me or any ides what to check.

Thanks

Log inor Register

to rate and post a comment

(23)
(1)

Abid Rehman

re: Siemens Sonoline G40

May 23, 2019 09:39

Please Replace the Cell of CPU.

Log inor Register

to rate and post a comment

(8)

gavin Sprong

re: re: Siemens Sonoline G40

May 24, 2019 08:54

Thank You..I will try that

Log inor Register

to rate and post a comment

(8)

gavin Sprong

re: re: Siemens Sonoline G40

May 30, 2019 09:14

Have replacedthe cell..its was dead. Fault still there.. OCP and AC led at the bottom are lit up ( these should all be off)

Log inor Register

to rate and post a comment

(8)

gavin Sprong

re: re: Siemens Sonoline G40

May 30, 2019 09:15

Do you have a service manual for this unit.OCP and AC led on PSU is on and these should be off

Log inor Register

to rate and post a comment

(1)

Tommy Agner

re: re: Siemens Sonoline G40

September 19, 2019 10:05

I am having the same problem. Can you tell me how to replace the cell of the cpu?

Log inor Register

to rate and post a comment

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment