(8)

javier andino

Siemens Xray Mobilett II Error 13

April 24, 2013 06:00

Hi ,
I have a mobilett II of siemens,
when i turn on the device only show me a Err 13
i have the service manual but i check all test and all are ok exect for the V1 on the D5 board somebody can say me what i should do ????

Log inor Register

to rate and post a comment

(40)
(4)

toni

re: Siemens Xray Mobilett II Error 13

April 24, 2013 09:42

Did you check D6 board?, , also have a fuse inside it, this board controls the filament, and if it's failing, the generator goes off. I hope it helps.

1 person liked this
Log inor Register

to rate and post a comment

(8)

javier andino

re: re: Siemens Xray Mobilett II Error 13

April 25, 2013 12:02

yes i chance this fuse and i have voltage to D5 but a i dont have voltages in the Tespoints of D5 board , are one signal missin thats signal come from D1 but i dont know why this signal is missing , what can i do ?
thanks

Log inor Register

to rate and post a comment

(40)
(4)

toni

re: re: re: Siemens Xray Mobilett II Error 13

April 26, 2013 09:42

If you have service manual, please check pages 7 and 25, and observe the TP on the board. my email is antolinrayosx@hotmail.com.

Log inor Register

to rate and post a comment

(12)

Waleed Ahmed

re: re: re: re: Siemens Xray Mobilett II Error 13

July 27, 2019 06:52

Hi evry body
Any body can help .. i have polymobil plus .. befor give me eeror 13 and now i have problem in trnsster BUP 314 bad whach one i can chane and whats problem with eeror 13 please any body can help

Log inor Register

to rate and post a comment

(1)

(2)

Bart Coppers

re: re: re: Siemens Xray Mobilett II Error 13

April 26, 2013 10:33

Could it be the fuse 125 amp 700 vac partnumber 9645532 ?

Log inor Register

to rate and post a comment

(30)
(7)
(2)

ARIAS RODRIGUEZ JOSE

re: re: re: re: Siemens Xray Mobilett II Error 13

April 26, 2013 12:39

Dear sir I have the error 12 I check the batt is ok.
anyone know how to solve this problem or how to test CMOS-RAM ?

Log inor Register

to rate and post a comment

(8)

javier andino

re: re: re: re: re: Siemens Xray Mobilett II Error 13

May 08, 2013 03:21

check the configuration of the switch i16 the correct configuration is close open open close open close close close

after this put the number 4 at open to go the the PRMODE and selec PR09 wait 5 secons and press the KV- 6 times at the 7 time press the two kv- and kv +
this work for me ªç!!!!!
i hope help you friend

Log inor Register

to rate and post a comment

(8)

javier andino

re: Siemens Xray Mobilett II Error 13

May 08, 2013 03:21

OK friends i found the reason of my error 13 , i chance the two ICs i1 and the i2 they are the lm723 from the d6 board and the v2 and v3 transistor, when i make the chance tha xray machine turn on ok withouth any error and the ready lamp are ok BUT when i do the first shot, the code error 13 appears again and the two transistor v2 and v3 and the two IC i1 and i2 are bad again i can i do ????
i check all components of the board d6 an the d5 and all are ok how can i know if my filament its ok????


Log inor Register

to rate and post a comment

(40)
(4)

toni

re: re: Siemens Xray Mobilett II Error 13

May 08, 2013 05:05

It seems to be a gassified tube, or one h.t. cable shotcircuited, but there are not h.t. cables, then look for in the h.t. head, there should be any abnormality inside, not the filament, but anything else, good luck.

1 person liked this
Log inor Register

to rate and post a comment

(8)

javier andino

re: re: re: Siemens Xray Mobilett II Error 13

May 13, 2013 04:44

hi toni checkin the HT i find sometime rare in this pictures you can se it http://imageshack.us/photo/my-images/211/itsnormalthis.jpg/
tell me is this mi problem ? the resistance should be more than 0.1 Ohm?
thanks for all

Log inor Register

to rate and post a comment

(88)
(9)

m sohail

re: re: re: Siemens Xray Mobilett II Error 13

August 18, 2013 12:05

hi sir i need mobilett II service manual can any help me msohail4007@yahoo.com

1 person liked this
Log inor Register

to rate and post a comment

Ramin gh

re: re: re: re: Siemens Xray Mobilett II Error 13

October 16, 2013 09:53

Hi Dear M.Sohail
Do you find mobilett II service manual i need this manual too .
could you please send that manual to me or any help me ? Thanks
raminghashami@yahoo.com

Log inor Register

to rate and post a comment

(7)

Kingsley Ekanem

re: re: Siemens Xray Mobilett II Error 13

August 03, 2017 01:50

Please I also have error 13 issue. I changed all the diodes v1 to v12. And the ICi1. It charges the capacitor bank and will trip off at about 180v then enter discharge mode automatically. Kindly advice on next action.

Log inor Register

to rate and post a comment

(12)

Waleed Ahmed

re: re: re: Siemens Xray Mobilett II Error 13

August 11, 2019 02:25

Please I also have error 13 issue. I changed all the trensstoir BUP 314 fuse 11 V and fuse 12 V It charges the capacitor bank and will trip off at about 180v then enter discharge mode automatically. Kindly advice on next action. Can you please help for this problem
Best regards Waleed

Log inor Register

to rate and post a comment

(88)
(9)

m sohail

re: Siemens Xray Mobilett II Error 13

August 18, 2013 12:05

hi sir i need mobilett II service manual any help me msohail4007@yahoo.com

Log inor Register

to rate and post a comment

mohammad Yousef

re: re: Siemens Xray Mobilett II Error 13

December 16, 2014 08:37

hi if you have the service please send to me my mail
Eng_mohd81@yahoo.com

Log inor Register

to rate and post a comment

(22)

joset-70

re: re: re: Siemens Xray Mobilett II Error 13

December 17, 2014 03:43

write me joseariasr@gmail.com

Log inor Register

to rate and post a comment

(12)

walid felah

re: re: re: re: Siemens Xray Mobilett II Error 13

May 23, 2015 07:58

can you kindly send me copy of SM my e mail libyamed@gmail.com

Log inor Register

to rate and post a comment

DMEDS

re: re: re: re: re: Siemens Xray Mobilett II Error 13

May 30, 2015 08:06

i have it contact me,

magmesme@gmail.com

regards

Log inor Register

to rate and post a comment

mohammad Yousef

re: Siemens Xray Mobilett II Error 13

December 16, 2014 08:37

hi I have the same problem we need the service manual can any one help me

Log inor Register

to rate and post a comment

mohammad Yousef

re: Siemens Xray Mobilett II Error 13

December 17, 2014 10:04

hi i need the service manual can any one send to me

Log inor Register

to rate and post a comment

(9)
(1)

Golden Nwobilor

re: Siemens Xray Mobilett II Error 13

November 17, 2016 01:09

Please I urgently need this seimens mobilett II x-ray service manual, if possible with the schematic block diagram this is my email. goldnwbi@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(7)

Kingsley Ekanem

re: Siemens Xray Mobilett II Error 13

August 13, 2019 10:43

Hi I have followed your discussion on err 13. I think your HV line needs to be checked at thr tube hrad fot arcking or low resistance. To confirm disconnect the hv line and try to expose.