Sonoline g40 no signal error message

Tags: Ultrasound

(3)

TD1989

Sonoline g40 no signal error message

September 19, 2019 10:04

Hi All,

I am having a problem with my sonoline g40. I am getting an error message saying that there is no signal from the signal cable. After a minute of the machine running it will shut down. Does anyone have any idea why it is doing this? This is a new machine to me/.

Log inor Register

to rate and post a comment

(90)
(15)

Juancho

re: Sonoline g40 no signal error message

September 20, 2019 02:25

Hi, TD1989
You should check the fan system and verify that the filters are clean. If the fault persists you should seek technical assistance.
Regards

Log inor Register

to rate and post a comment

(3)

TD1989

re: re: Sonoline g40 no signal error message

September 23, 2019 08:58

Do you have any kind of details on how to do that? or someone I can reach out to for technical assistance? I am completely new to working on these things, just trying to get it up and running. The machine was in a working office in december, and has sat there since the office was shut down. There cant be much wrong with it.

Log inor Register

to rate and post a comment

(9)

AbidRehman

re: Sonoline g40 no signal error message

September 23, 2019 09:17

Also check Cell voltage on CPU Board

Log inor Register

to rate and post a comment

(3)

TD1989

re: re: Sonoline g40 no signal error message

September 23, 2019 04:28

Do you by chance have the software to reinstall the program? my computer guy is gonna take a look at it and says we will more than likely need the software install disc to get this thing back up and running.

Log inor Register

to rate and post a comment

(90)
(15)

Juancho

re: re: re: Sonoline g40 no signal error message

September 24, 2019 09:47

Hi, TD1989
I think that is a problem of over temperature. Check fan & filter.

Regards

Log inor Register

to rate and post a comment