> Startup Problem on Asteion Super 4

Startup Problem on Asteion Super 4

(4)

Sigid Subagyo

Startup Problem on Asteion Super 4

July 31, 2017 10:26

Dear Sir's..,
My name Sigid from Jakarta,
I have start-up problem on Asteion S4..
It says "Can not re-start RTM" and I got these messages from RTM terminal :
Star Interface.c. 79 tStart : Board in slot 18 must be G4-DTB board
StartupTask.c 82 tStart : Error in StarPCIInit
A day before, it was normal..
I only have service manual and software volume manual, unfortunately, there is no procedure guide to set the RTM.
Thank you very much if you could spare your time for me..


Sincerely your's
Sigid Subagyo
ssubagyo@yahoo.com

Log inor Register

to rate and post a comment

(11)
(3)

Vicente Sarate

re: Startup Problem on Asteion Super 4

July 31, 2017 05:38

Hi Siguid.

It seems you have trouble with a MHR -DP board 18 slot. Please swap slot 18 MHR-DP with slot 17 also MHR-DP. Now you can see if the message on RTM Terminal change to board in slot 17. If this happen, then the defective part is the MRH-DP slot 18. Please tell me if you can do that.

Also you can contact me at martin.verdun@focus-mg.com

Log inor Register

to rate and post a comment

(4)

Sigid Subagyo

re: re: Startup Problem on Asteion Super 4

August 11, 2017 07:56

Thank you for your reply Martin,
I already switched location between two MHR-DP, but the problems remains the same..

Regard's
Sigid Subagyo.

Log inor Register

to rate and post a comment

Infusion Pump Repair Corp, Certified Expert Repair

Infusion Pump Repair (IPR) is a certified medical support and solutions company. We specialize in the repair, refurbishment and resale of infusion pumps. Call 1-855-477-8866 or email us: info@infusionpumprepair.com DOTmed Certified

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED