I have a new RF AMP (20KW) for AN8119H. Because there is a AN8119T RF AMP burned out. I want to replace AN8119T with AN8119H. Just now I replaced AN8119T with AN8119H on a Toshiba 1.5T MR system. The situation is this, When using the knee coil imaging, AN8119H RF run normal and imaging good (knee coil spontaneous self collection, low power), in the use of other coils, AN8119H does not work. In manual, the 25PIN RS422 interface that AN8119H is different from AN8119T.
Who has the method to technically change AN8119H to fit AN8119T?.

Log inor Register

to rate and post a comment

(146)
(17)

Wang Zhixuan

re: Toshiba 1.5T MR RF AMP repaire

August 13, 2017 02:08

You need to check what different in AN8119T to AN8119H,first change the wiring in the RS422 connector.