Toshiba data transfer system in the gantry is abnormal

Tags: CT, TOSHIBA

(18)
(5)

Renato Carpio

Toshiba data transfer system in the gantry is abnormal

February 23, 2023 10:26

My appreciation to anyone that take the time on this reading, have Toshiba Activion 16 CT scanner, very simmilar than Aquilion 16 CT, get this yellow message error at start.
Data transfer system in the gantry is abnormal
there are 3 buttoms to select, I choise details to show:
---ERROR_INF dum---
SSADI STATUS
ss_state(0x0000), valid_bit(0x0
GTS STATUS
mode(OR), exec(RR), error(0x008
interlock Slide(03), height (00)
free&light(00), out2censor cens
MUDAT status(0x0000)
XC STATUS
mode(OR), exec(RR), error(0x103
---ADD ERROR INF DUMP---
addinf
init task: mudat check error

There is no change when I click on ERROR CLEAR, Im not trained technician, I fix this scanner several times with RTM box issues, but its the first time that I get gantry errors, so, Im new on this, any recomendation, service or troubleshooter document will be very helpfull. Ofcourse, I can fallow specific instructions to fix, change or clean.

Appreciate in advance any help.
Best regards

Log inor Register

to rate and post a comment

(45)
(9)
(4)

Robert Nurbekyan

re: Toshiba data transfer system in the gantry is abnormal

February 24, 2023 10:57

Hello Renato,
First of all, check Power supplies in rotation part of gantry.

Take care.

Log inor Register

to rate and post a comment

(217)
(14)
(8)

Carlos Duran

re: Toshiba data transfer system in the gantry is abnormal

February 24, 2023 10:58

Hello Renato,

Looks like a different issue, but console see that. Need to check PCB's inside gantry, to isolate what happens.

Log inor Register

to rate and post a comment

(18)
(5)

Renato Carpio

re: re: Toshiba data transfer system in the gantry is abnormal

February 27, 2023 01:10

Hi !! thanks for repply, I already check PSUs in rotary and stationary sides, every think is ok, i find red leds on GCIFA board, check on service manual, there are 89 - 90 and 74 leds numbers. It mean some problem about transfer data betwen stationary and rotary side. 74 led mean FIFO (memory) is empty for channels 7 and 8. Have a friend that see birlogs and he tell me there is a problem on SPD6 or SPD7, there are the channels that read on stationary from rotary side. Im stuck here, he tell me this is a very dificult about photo diodes. So, if some body have documentation about how to troubleshot this area ( becasue problem can come from OPCONTA ) or about graphic MUDAT mantenance will be great. I find some service manual about that but this is only flux diagram, I need to know how to do in good way, I dont want to broke some thing more.

Log inor Register

to rate and post a comment

(18)
(5)

Renato Carpio

re: re: re: Toshiba data transfer system in the gantry is abnormal

March 13, 2023 10:20

Please, any recomendation about this will be wellcome, Im stuck with this and scanner is stoped more than a month, until now, no PSUs are bad, OPCONTA or GCIFA good, allways same message: The data transfer system in the gantry is abnormal.
Code (0xb7100001) dev(0x00010007) Pno(8004)

Log inor Register

to rate and post a comment

(4)

Rick1987

re: re: re: re: Toshiba data transfer system in the gantry is abnormal

March 15, 2023 11:58

Try to clean Mudat PWDs with dry soft cloth and vaccum cleaner with a brush

Log inor Register

to rate and post a comment

(13)
(1)

fouts

re: re: re: re: re: Toshiba data transfer system in the gantry is abnormal

March 27, 2023 10:40

data xsfer is thru optical sensors. get strong vacuum to 'blow' all gantry and rear of gantry rotation section,little dust causes big problems.

Log inor Register

to rate and post a comment

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com Call 813-925-1314

DOTmed Certified DOTmed 100
advertisement

Upcoming Auction Events

Surplus Service
TIMED Ending Soon
Mar 27, 2023 (71 lots)
Tekyard Medical
TIMED
Mar 29, 2023 (214 lots)
Therapeutic Technologies
TIMED
Mar 30, 2023 (1553 lots)
ZRG Medical
TIMED
Apr 07, 2023 (78 lots)