Hi every one
I need help for some problem due to Toshiba x-ray machine model KXO15R
I would be gratefl if any one got well experience with that model contact with me urgently.
Thanks in advance
Alaa Mohammed

Log inor Register

to rate and post a comment

(10)
(1)

WellingtonTrajano

re: Toshiba KXO 15R

April 17, 2017 10:18


Hi friend, I already did some work on this model KXO15R whats your question

1 person liked this
Log inor Register

to rate and post a comment

(21)
(1)

Alaa Mohamed

re: re: Toshiba KXO 15R

April 18, 2017 10:29

Thanks..by turning on the machine it gives Overload without alarms, by pressing restart button and set key it gives error 90..I checked the whole filament circuit and its OK, I checked all flow charts for error 90 in the performance book and all are OK, I checked the FH Pwb and its OK.
Its mentioned in the performance book that the tube head might be gaseous,although I think its not the reason but Im going to try with another tubehead..I think itsnt the reason because the error appears by turning on the machine;not by taking the exposure..
I would be grateful if you could help
Regards
alalateef@yahoo.com

Log inor Register

to rate and post a comment

(21)
(1)

Alaa Mohamed

re: re: Toshiba KXO 15R

April 18, 2017 10:29

Thanks
Pls send your email
Thanks again

Log inor Register

to rate and post a comment

(21)
(1)

Alaa Mohamed

re: re: Toshiba KXO 15R

April 18, 2017 11:01

Pls send your email to be in contact.
Thanks

Log inor Register

to rate and post a comment

(10)
(1)

WellingtonTrajano

re: re: re: Toshiba KXO 15R

April 19, 2017 08:19

tom.jp.pb@gmail.com

Log inor Register

to rate and post a comment

(21)
(1)

Alaa Mohamed

re: re: re: re: Toshiba KXO 15R

April 20, 2017 10:21

I sent you an email...uf you havent recieved it pls send your comment through dotmed or to alalateef@yahoo.com
Thanks alot

1 person liked this
Log inor Register

to rate and post a comment

(10)
(1)

WellingtonTrajano

re: re: re: re: re: Toshiba KXO 15R

April 23, 2017 10:25


Hello friend you already checked the transformer and I text as follows with the help of a dazzling 60watts bulb you connect one of the wires of the bulb in the FC and the other end of the wire in the FF or FG and makes an incidence to see if the bulb lights

Log inor Register

to rate and post a comment

(21)
(1)

Alaa Mohamed

re: re: re: re: re: re: Toshiba KXO 15R

April 25, 2017 08:19

Thanks for your kind reply.
I did check the whole filament circuit by energising the CNNE terminals by an external power supply...the filament lighted up...and checked the FH Board with a working machine anf its OK

Log inor Register

to rate and post a comment

RaySafe helps you avoid unnecessary radiation

RaySafe solutions are designed to minimize the need for user interaction, bringing unprecedented simplicity & usability to the X-ray room. We're committed to establishing a radiation safety culture wherever technicians & medical staff encounter radiation.

 

 

Advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED