Alaa Mohamed

Toshiba KXO15R

April 20, 2017 10:22

By turning on the machine (Toshiba KXO15R)it gives Overload without any alarms, by pressing restart button and set key it gives error 90..I checked the whole filament circuit and its OK, I checked all flow charts for error 90 in the performance book and all are OK, I checked the FH Pwb (with another working machine)and its OK.
Its mentioned in the performance book that the tube head might be gaseous,although I think its not the reason but Im going to try with another tubehead..I think itsnt the reason because the error appears by turning on the machine;not by taking the exposure..
I would be grateful if any colleague could help or give some advice
Regards
manalgooda@yahoo.com

Log inor Register

to rate and post a comment

(2195)
(391)
(11)

Lenny Place

re: Toshiba KXO15R

April 21, 2017 10:36

I do not have a manual for this model but had a similar error like this on a KXO80 system and it turned out that I lost the battery backup and I lost the configurations and calibrations. On the kxo80 I was able to restore the defaults using the switches in the console, after the defaults were loaded I was able to make an expose.

1 person liked this
Log inor Register

to rate and post a comment

(19)
(1)

Alaa Mohamed

re: re: Toshiba KXO15R

April 23, 2017 10:24

Thanks alot for your reply..
Yes I found a faulty battery in the Control Board..I replaced it but the manual doesnt mention any sequence for reloading any data..
Thanks again

Log inor Register

to rate and post a comment

(2195)
(391)
(11)

Lenny Place

re: re: re: Toshiba KXO15R

April 24, 2017 10:15

Other models kxo 22 and 32r it is done by selecting the following conditions for initializing
SR2 1 SRl 1 Contents
0 0 I I The 7-segment LED on the CPU 22R PWB is counted up in
l-second increments.
0 1 1 1 Heat capacity of tube 1 is displayed on the panel.
0 2 Heat capacity of tube 2 is displayed on the panel.
4 1 0 IAll LEDs on the panel are lit.
C I 0 I Setting of the X-ray output data table
c 1 Setting of the system condition setting data table
C I 2 I Initialization of the X-ray output data table
c 3
I I
Initialization of the system condition setting data
table
c 4 Initialization of the radiographic condition data table
c 5 Initialization of the error table
D 0
I I
The watchdog timer is stopped when the registration
switch is pressed.
D 1 The tube current maximum value is halved when the
registration switch is pressed.
D 2 The X-ray exposure timer is stopped when the
registration switch is pressed.

Log inor Register

to rate and post a comment

(19)
(1)

Alaa Mohamed

re: re: re: re: Toshiba KXO15R

April 25, 2017 08:19

Thanks alot..
Its completely different..
Thanks again

Log inor Register

to rate and post a comment

Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students. DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED