Unable to ping Fuji XG-1 CR

September 03, 2020 10:54

Hi all, I am unable to ping Fuji XG-1 CR anymore. The CR boots without issues but I am unable to ping it from the workstation. I changed the CPU board and I was able to ping, so I could narrow the problem to the CPU board. The led light on the network port is on as described in the service manual but it still can't ping. I am thinking of installing the CR software but I can't do this since it is not connected to the network. I tried connecting to the CPU board through the serial cable via Putty. It worked and showed Fuji intro but I can't do anything else.

Pls, I need help to restore the network back to the CPU board. Any help will be appreciated.


Kind regards.

Log inor Register

to rate and post a comment

(12)

Amir Jemal

re: Unable to ping Fuji XG-1 CR

September 07, 2020 11:16

I will help you, no problem
1.Check the setting IP address in RU and CL
2. Check your getaway setting in RU and CL
3. Try to plug the network cable directly to the PC, remove the Network switch.
4, finally, Install RU software , hope it will fix your system.
Good luck
Amir

Log inor Register

to rate and post a comment

(363)
(73)
(2)

juan pablo amado

re: Unable to ping Fuji XG-1 CR

September 14, 2020 10:47

Have you tried using native network addresses?
172.16.1.10 (RU)
172.16.1.20( PC)