Vantage G-amp time out

October 09, 2018 04:37

Hi. I need help with a Toshiba MRI Exelart Vantage MRT-1503. When system is booting up, gradient amps turns on by the 400v contactor. But the HV never turns on. It just keep turning off and turning on 400v contactor again and again. And host CPU just say gradient amp error. Time out. I already test CANV1 board, D16 and D17 board on another similar MRI system and it works.

Log inor Register

to rate and post a comment

(2)

Marcus Stephens

re: Vantage G-amp time out

October 11, 2018 08:08

Wish I could give you some tech support on this, I usually deal with GE. If you need a part repaired on this let me know I can repair it.