XC error Toshiba auklet CT scan

March 02, 2014 08:44

overlong exposure time error in the xc, Reset xc and execute the command again.
[xc reset] [detail] [ok]
when you click details you get the following details:
bno code error
07 E13 abnormal arcing was detected in the high-volta
24 E0 XC error
27 E3 Exposure time error

Log inor Register

to rate and post a comment

(141)
(20)

Wang Zhixuan

re: XC error Toshiba auklet CT scan

March 03, 2014 08:57

According you said,there is arcing in your CT scanner,how long is your tube used? Another cause may be HV generator.Could you check them by yourself?If you cann't,please contact me.My email is 362689810@qq.com.

Best regards.

Mr.Wang zhixuan

Log inor Register

to rate and post a comment

(3)

Mohamed Khalifa

re: re: XC error Toshiba auklet CT scan

March 15, 2019 03:37

Dear wang.
How are you, I hope you are doing fine.
I have error E27
Would you please send soft copy of Auklet service manual.

Best regards.
Mohamed Khalifa.