> Ziehm Exoscop 7000 Error 26 during exposure - service manual

Ziehm Exoscop 7000 Error 26 during exposure - service manual

Todd Starnes

Ziehm Exoscop 7000 Error 26 during exposure - service manual

January 11, 2011 12:40

Good morning. I need a service manual for a ziehm exoscop 7000 or some insight as to what the error code E26 represents when trying to make an exposure. I can make a 50kv manual exposure?

Is it tube arc??
Any assistance is greatly appreciated.
Thanks,

Todd Starnes

Log inor Register

to rate and post a comment

(49)
(6)
(32)

John Mayes

re: Ziehm Exoscop 7000 Error 26 during exposure - service manual

July 12, 2012 01:46

Did you find out what the error 26 was? Im having the same problem?

Log inor Register

to rate and post a comment

(49)
(6)
(32)

John Mayes

re: re: Ziehm Exoscop 7000 Error 26 during exposure - service manual

November 26, 2012 12:34

No, never found out, we left it unplugged for sometime, plugged it back in and now reading error 32, which I was told means batteries on cpu board need replacing, we decided to sale as parts unit

Log inor Register

to rate and post a comment

(102)
(14)

Alaa Mohammed

re: Ziehm Exoscop 7000 Error 26 during exposure - service manual

February 27, 2013 04:43

John LoPresti
re: re: re: Ziehm exposcop 7000 compact error code 26
July 25, 2012 08:00
It means there was a generator error, could be the generator needs to be recalibrated, or a bad connection to the tube or bad wire withing the cable harness or worse case a faulty, arcing tube. Errors 26 & 29 in most cases means the tube is arcing and needs to be replaced but you could get lucky with a recalibration or checking the connections to the tube. The high voltage is developed at the tube head which is usually at fault...good luck

Log inor Register

to rate and post a comment

OiHealthcare - ISO 9001 & 13485 Certified - Call 1-888-278-9933

Oxford Instruments Healthcare, is a leading ISO 9001 & 13485-certified organization, that specializes in providing quality refurbished CT & MRI systems, service, mobile units & parts at prices you can afford. Call 888-278-9933 Email DOTmedinfo@oxinst.com

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED