Ziehm Solo And Vision X Ray Problem - Service Technicians Forum

Over 1050 Total Lots Up For Auction at Three Locations - OK 05/12, CA 05/14, NJ 05/24

Ziehm solo and vision X ray problem

Tags: C-Arm, ZIEHM

Posted by an Anonymous User

Ziehm solo and vision X ray problem

April 29, 2021 11:37

Dear All,,
We have use Ziehm solo and vision devices, but this devices same problem,,

The device is powering up normally
does not give an error or warning
setting operating modes
x ray foot pedal is pressed
the device is turning on normally
does not give an error or warning
setting operating modes
x ray foot pedal is pressed
The x ray light turns on for 1-2 seconds, first a poor quality image comes, then the screen turns black, the fluoroscopy stops the x ray, the device does not give an error code
sometimes warning 1066 is on the screen,

Can you share your thoughts with me,

Best regards.
Muzaffer....


Log inor Register

to rate and post a comment

(19)
(4)

walter molina

re: Ziehm solo and vision X ray problem

April 30, 2021 09:54

Hello may be you have a problem with the XRay tube.

Log inor Register

to rate and post a comment

(7)

(1)

MUZAFFER OZMEN

re: re: Ziehm solo and vision X ray problem

May 03, 2021 12:54

Hello,

First of all, thank you for your answers and attention ,,

I looked at the log files (zivi.log etc.) of the devices, the x ray model, which is VISION MODEL, gives tube error. I opened the cap with the tube, saw oil foaming out of the tube box .. here, there is a serious problem with the x ray tube ..

Z─░EHM gives 1066 can bus error in SOLO version, However, I checked all cards and RJ cables,
However, when I press the x ray button, kv values on lcd screen are maximized. However, it does not give x rays ,,
Voltages to the x ray tube are normal, I checked the cards. The tube is probably broken, it does not respond to the incoming data ..

I will share the developments

Many thanks

Log inor Register

to rate and post a comment

(3)

jagalindof

re: re: re: Ziehm solo and vision X ray problem

May 04, 2021 11:26

Hi,
Definitely you have serious problems in both units.
In Solo unit you can extract logs in the same way as in Vision.
Remember that you will need an usb service stick in order to enter the complete service menu, adjust parameters and more. Otherwise, you couldn't repair the unit.
This usb service stick is provided by Ziehm only to their certified engineers.

Log inor Register

to rate and post a comment

(163)
(22)

Tibor Ivanszky

re: Ziehm solo and vision X ray problem

April 30, 2021 09:56

Examine the connectors of the interconnecting cables between monitor and stand throghly . Check every pin missing or bent.

Log inor Register

to rate and post a comment

(3)

jagalindof

re: Ziehm solo and vision X ray problem

April 30, 2021 09:57

Hi Muzaffer.
To begin with, Solo and Vision are two different units. Which one is the unit with problems?
What you explain is too general to diagnose the issue.
1066 is an internal communication error, maybe because there is a faulty board/module.
You will need a "report package" where you can find the errors log file. You can generate the files pluging in an usb stick and pushing the touch screen up/right (r-ray symbol) on the screen, once the unit is powered up. It will ask you about generate the report package.
Inside you will find some encripted files and the error logs, where you can find more detailed info about your problem.

Log inor Register

to rate and post a comment

C-Arm Solutions for Every Need...Service, Sales & Rentals

Industry leader with 25+ years offering customers New and Refurbished C-Arms options including long term service agreements. National Coverage for equipment delivery and service. Contact our team at 1.800.528.7616 or email at t.orlando@irad.net

DOTmed Certified DOTmed 100
advertisement

 

Hitachi

 

RSTI Newsletter

Upcoming Auction Events

Therapeutic Technologies
TIMED Ending Soon
May 07, 2021 (177 lots)
Premier Medical Depot
TIMED
May 12, 2021 (595 lots)
ZRG LLC.
TIMED
May 14, 2021 (162 lots)
Surgamed
TIMED
May 24, 2021 (329 lots)