Over 40 Florida Auctions End Tomorrow 09/28 - Bid Now
Over 2250 Total Lots Up For Auction at Four Locations - OK 09/29, NJ Cleansweep 09/30, CA 10/01, NJ 10/05

Ziehm Solo C-Arm

(16)
(5)

Hasan sevinc

Ziehm Solo C-Arm

September 07, 2021 02:28

Ziehm in solo scope device
I get error E1719 Gain adjustment not possible. I would appreciate it if you could help me understand what it means.

Log inor Register

to rate and post a comment

(4)
(1)

jagalindof

re: Ziehm Solo C-Arm

September 08, 2021 10:59

Hi Hasan, the video-signal from the CCD-camera is missing. Check cabling and CCD suply.

1 person liked this
Log inor Register

to rate and post a comment

(16)
(5)

Hasan sevinc

re: re: Ziehm Solo C-Arm

September 14, 2021 11:45

Thank you very much Jegalindof problem solved. The video cable is plugged in the wrong place. But now it gives another error.
E1226: CAN modules missing: 17.9
Do you know what this means?

Log inor Register

to rate and post a comment