Ziehm Vision Fd Vario 3d - Service Technicians Forum

Over 1400 Total Lots Up For Auction at Six Locations - MD 06/16, UT 06/18, France 06/21, PA 06/22, UT 06/24, CA 06/25

Ziehm vision fd vario 3d

(1)

Mikhail Zhigalkin

Ziehm vision fd vario 3d

March 05, 2021 02:56

Help me please, need service manual for Ziehm vision fd vario 3d, error code:e1015, e1017, e1070, I will be grateful for any help. Thanks.

Log inor Register

to rate and post a comment

(9)
(4)

Dawid Haraƛ

re: Ziehm vision fd vario 3d

March 08, 2021 10:53

Hi Mikhail,

E1015 - wrong machine data version detected during power-up
E1017 - Test of "FAULT" signal detected
E1070 - Temperature sensor 2 value beyond permissible range

I have service manual, contact me

Log inor Register

to rate and post a comment

(28)
(3)

Igor69

re: re: Ziehm vision fd vario 3d

April 05, 2021 10:48

Err 1070 is Differenz between Tempsensor 1 and Tempsensor 2. I had same error, Sensor 2 was defective. The sensor is in the tank. You can check at TP for Tempsensor 1 and 2, they are labled. You can check cold and with increment temp, they should be nearly equal. Sensor is PT1000, you can test with a sensor pt1000 with temp-coupling on the tank. Regards Igor

Log inor Register

to rate and post a comment

(25)
(2)

sergio caballero

re: re: re: Ziehm vision fd vario 3d

June 02, 2021 02:12

Thanks for your information, it was very useful and helped me solve the error code 1070

Log inor Register

to rate and post a comment

(25)
(2)

sergio caballero

re: re: Ziehm vision fd vario 3d

May 31, 2021 01:40

Hello, could you help me with ziehm vision service manual, because I have problem with error code 1070, thanks
Whith best regards

Log inor Register

to rate and post a comment

(66)
(8)

Holger Fritze

re: re: re: Ziehm vision fd vario 3d

June 01, 2021 10:24

Hello Sergio, i dont have the manual, but you only have to check this: Err 1070 is Differenz between Tempsensor 1 and Tempsensor 2. We had same error, Sensor 2 was defective. The sensor is in the tank. You can check at TP for Tempsensor 1 and 2, they are clearly labled! You can check cold and with increment temp, they should be nearly equal. Sensor is PT1000, you can test with a sensor pt1000 with temp-coupling on the tank. Best regards Holger

Log inor Register

to rate and post a comment

(66)
(8)

Holger Fritze

re: Ziehm vision fd vario 3d

June 01, 2021 10:24

look at the post April 05, 2021 10:48

Log inor Register

to rate and post a comment

(25)
(2)

sergio caballero

re: re: Ziehm vision fd vario 3d

June 02, 2021 02:12

thaks i have solve the problem

Log inor Register

to rate and post a comment

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com Call 813-925-1314

DOTmed Certified DOTmed 100
advertisement

 

Hitachi

 

RSTI Newsletter

Upcoming Auction Events

Tekyard Medical
TIMED
Jun 16, 2021 (289 lots)
Primis Medical
TIMED
Jun 18, 2021 (122 lots)
Alternup Medical
TIMED
Jun 21, 2021 (143 lots)
Medical Equipment Superstore
TIMED
Jun 22, 2021 (570 lots)
Primis Medical
TIMED
Jun 24, 2021 (221 lots)
ZRG LLC.
TIMED
Jun 25, 2021 (101 lots)