fuji drypix 4000 or 7000 IP ADDRESS

August 21, 2012 09:00

I need to change the IP ADDRESS of our fuji drypix 4000 and 7000.anyone can help me how? or willing to share the service manual...thanks in advance.

1 person liked this
Log inor Register

to rate and post a comment

(77)
(13)
(2)

Gery Rodriguez

re: fuji drypix 4000 or 7000 IP ADDRESS

August 21, 2012 06:41

in order to change IP addresses on any DP4000 and 7000 you need the DI-Tool software CD or program to connect to it. You cannot do this from the LED screen. The service manual tells you how but no DI-Tool, no luck.

1 person liked this
Log inor Register

to rate and post a comment

(55)
(9)

ali hashemyan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 09, 2013 06:48

hi ,
do u have service manual of drypix 7000.
can u send it to me?
my email:
ali45922002@yahoo.com

1 person liked this
Log inor Register

to rate and post a comment

(75)
(7)
(2)

hemrajsinh chauhan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 29, 2016 03:13

i need this pls help me

Log inor Register

to rate and post a comment

(1)
(1)

Poka Tikili

re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 27, 2018 09:43

I am trying to connect fujifilm drypix 4000 to my laptop, so that I can print from dicom files, well where can I get DI-Tool?

1 person liked this
Log inor Register

to rate and post a comment

(147)
(22)

Wang Zhixuan

re: fuji drypix 4000 or 7000 IP ADDRESS

August 22, 2012 12:57

Yes, you need DI-Tool software to change the IP address, if you cann't get the software,I suggest you change the IP address in master,such as CT,MRI,DR etc.

1 person liked this
Log inor Register

to rate and post a comment

(55)
(9)

ali hashemyan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 09, 2013 06:48

hi ,
do u have service manual of drypix 7000.
can u send it to me?
my email:
ali45922002@yahoo.com

Log inor Register

to rate and post a comment

(55)
(9)

ali hashemyan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 09, 2013 06:48

hi ,
do u have service manual of drypix 7000.
can u send it to me?
my email:
ali45922002@yahoo.com

Log inor Register

to rate and post a comment

(6)

Alexandros Palaiologos

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

February 29, 2016 02:18

Ηello! ,
Do you have the service manual of DryPix 7000.
Can you send it to me?

my email: alexpal83@hotmail.com

Thanks,
Alexandros

Log inor Register

to rate and post a comment

(7)
(1)

Shahriar Shokri

re: re: fuji drypix 4000 or 7000 IP ADDRESS

September 08, 2013 04:17

Dear Wang
Can you send me a DI-tools via Email?

Log inor Register

to rate and post a comment

(30)
(5)

biomedical

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

September 09, 2013 06:28

Can you send me a DI-tools via Email ZAINMRI1@GMAIL.COM

Log inor Register

to rate and post a comment

(47)
(1)

Deniss Stalnov

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 13, 2017 10:23

I have the same problem. Need help. Could you please send me Di-Tools too. denis.stalnov@gmail.com

Log inor Register

to rate and post a comment

(47)
(1)

Deniss Stalnov

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 13, 2017 10:23

I have the same problem. Need help. Could you please send me Di-Tools too. denis.stalnov@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(75)
(7)
(2)

hemrajsinh chauhan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 29, 2016 03:13

pls send me link of this software raj_meditech@yahoo.co.in

Log inor Register

to rate and post a comment

(88)
(9)
(1)

Arbjan Rusi

re: fuji drypix 4000 or 7000 IP ADDRESS

August 22, 2012 11:00

You have to enter at M-uttility then system managment, network and then you change the ip address.this is valid for drypix 7000a.you don't need DI tool for changing IP address.

Log inor Register

to rate and post a comment

(5)
(2)

mar sison

re: re: fuji drypix 4000 or 7000 IP ADDRESS

August 25, 2012 11:40

hi can you share to me the service manual..i need the step by step procedure.thank you.

Log inor Register

to rate and post a comment

(88)
(9)
(1)

Arbjan Rusi

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

August 28, 2012 10:29

give me your email and I will send to you the procedure to enter to the M-utility. it is very simple

Log inor Register

to rate and post a comment

(3)

Lasaynu Latimaha

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

October 11, 2012 12:51

leshxx@yahoo.com

Log inor Register

to rate and post a comment

(1)

José Héctor Ríos Navarro

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 24, 2013 10:17

jhrios@hotmail.com

Log inor Register

to rate and post a comment

(55)
(9)

ali hashemyan

re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 09, 2013 11:55

hi ,
do u have service manual of drypix 7000.
can u send it to me?
my email:
ali45922002@yahoo.com

Log inor Register

to rate and post a comment

(30)
(5)

biomedical

re: re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

September 09, 2013 06:28

PLZ GIVE ME DI TOOLS SOFTWEAR

1 person liked this
Log inor Register

to rate and post a comment

(47)
(1)

Deniss Stalnov

re: re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 13, 2017 10:23

I have the same problem. Need help. Could you please send me Di-Tools too. denis.stalnov@gmail.com

Log inor Register

to rate and post a comment

Ingelectra

re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

January 11, 2015 12:33

dencalsa@yahoo.com.mx Please send me to my too

Log inor Register

to rate and post a comment

(55)
(9)

ali hashemyan

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 09, 2013 11:55

hi ,
do u have service manual of drypix 7000.
can u send it to me?
my email:
ali45922002@yahoo.com

Log inor Register

to rate and post a comment

(47)
(7)
(6)

KONSTANTINOS PATTICHIS

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

November 04, 2013 01:27

My email info@medisofpc.gr
Send me also the procedure
Thanks Kostas

Log inor Register

to rate and post a comment

(2)

(54)

Luke Monteleone

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

November 12, 2015 06:14

Hi, looking to enter new IP address information and AE titles. Can you help me?

Log inor Register

to rate and post a comment

(75)
(7)
(2)

hemrajsinh chauhan

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 17, 2016 05:13

raj_meditech@yahoo.co.in pls send me information

Log inor Register

to rate and post a comment

(5)

Imran Latif

re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 17, 2013 05:05

hi

Log inor Register

to rate and post a comment

(55)
(9)

ali hashemyan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 09, 2013 06:48

hi ,
do u have service manual of drypix 7000.
can u send it to me?
my email:
ali45922002@yahoo.com

Log inor Register

to rate and post a comment

(5)

Haytham Bannoura

re: fuji drypix 4000 or 7000 IP ADDRESS

August 22, 2012 05:14

you can do that by entering the M-Utility. you have to press on the upper left corner then after and immediately on utility then select system management then network then you can change the IP address of the printer.

Log inor Register

to rate and post a comment

(5)
(2)

mar sison

re: re: fuji drypix 4000 or 7000 IP ADDRESS

August 25, 2012 11:40

hi i tried but its not work to enter M-utility..how long do i need to press the upper left corner?and after that is it the UTILITY button that i need to press? can you give me a detailed procedure to enter M-utility..is it applicable for both fuji drypix 4000 and 7000? thank you for reply.you can email me at bmedperps@gmail.com

Log inor Register

to rate and post a comment

(39)
(6)
(1)

Raymundo Miranda

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

September 03, 2012 11:51

I send you a brief manual to change teh ip adress in your mail

2 people liked this
Log inor Register

to rate and post a comment

Shawn Broussard

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

October 13, 2012 01:09

You need DI Tool to change AE Title not IP Address

Log inor Register

to rate and post a comment

(5)

Imran Latif

re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 17, 2013 05:05

i need a Di-Tool any body help me.

Log inor Register

to rate and post a comment

(2)

ali hoseini

re: re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

May 29, 2013 08:08

after ready printer , u must peress on led screen (first on "ready" and on "Utility" button)and u can enter to Mutlity.You have to enter at M-uttility then system managment, network and then you change the ip address.this is valid for drypix 7000a.you don't need DI tool for changing IP address

Log inor Register

to rate and post a comment

(1)

Jose Almada Teixeira

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

September 12, 2013 07:07

have service manual of drypix 7000.
can u send it to me?
my email: almadateixeira@gmail.com

Log inor Register

to rate and post a comment

(77)
(13)
(2)

Gery Rodriguez

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 29, 2013 02:29

once again . . . this only works best with DI-Tool Software.

Log inor Register

to rate and post a comment

(30)
(5)

biomedical

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

January 12, 2015 10:10

Gery plz give me di tools softwear

Log inor Register

to rate and post a comment

(47)
(1)

Deniss Stalnov

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 13, 2017 10:23

I have the same problem. Need help. Could you please send me Di-Tools too. denis.stalnov@gmail.com

Log inor Register

to rate and post a comment

(8)
(1)

chea chandy

re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 20, 2020 11:16

Hello Sir, Could you tell me how to add new client to my Drypix 7000. I try to put AE title, IP address and Port of the printer on my new console but It cannot send the image to the printer for printing. Could you tell me how to do it?

Thank you in advance

PS: I have DI-Tool on my computer.

Log inor Register

to rate and post a comment

(2)

ali hoseini

re: fuji drypix 4000 or 7000 IP ADDRESS

May 29, 2013 08:08

after printer is ready press on led screen(first on "ready"and press Utility button)
You have to enter at M-uttility then system managment, network and then you change the ip address.this is valid for drypix 7000a.you don't need DI tool for changing IP address

Log inor Register

to rate and post a comment

Yuriy Senelshchikov

re: re: fuji drypix 4000 or 7000 IP ADDRESS

January 12, 2015 04:43

It is work with FM-DP L?

Log inor Register

to rate and post a comment

(1)
(1)

Ken Thomas

re: fuji drypix 4000 or 7000 IP ADDRESS

September 12, 2014 09:56

We have a drypix 4000 and I also need instructions on modifying the IP config. I have the DI-Tool but do not know how to use it. I have attempted to access the settings at the printers LCD console but cant seem to find any way to update IP config there either. Can anyone here be of assistance?

1 person liked this
Log inor Register

to rate and post a comment

(30)
(5)

biomedical

re: re: fuji drypix 4000 or 7000 IP ADDRESS

January 13, 2015 10:18

Can u give me my email zainmri1@gmail.com

Log inor Register

to rate and post a comment

(2)

(54)

Luke Monteleone

re: re: fuji drypix 4000 or 7000 IP ADDRESS

November 12, 2015 06:14

Did you ever find out how to do this?

Log inor Register

to rate and post a comment

(75)
(7)
(2)

hemrajsinh chauhan

re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 17, 2016 05:13

i need di tool pls send me on raj_meditech@yahoo.co.in

Log inor Register

to rate and post a comment

Marioo Leone

re: fuji drypix 4000 or 7000 IP ADDRESS

June 12, 2015 07:41

Could you please send me the DiTool soft? mawasz@wp.pl

Thanks!

Log inor Register

to rate and post a comment

Douglas Lobo

re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 22, 2015 11:33

Could you please send the soffware to me?
suporteti@amelimeira.unicamp.br

Log inor Register

to rate and post a comment

(41)
(12)

Luis PASCUAL

re: fuji drypix 4000 or 7000 IP ADDRESS

March 04, 2016 08:52


Hello! I can help you. E-mail me at crisjoseq@hotmail.com

Log inor Register

to rate and post a comment

Bonnato Bonis

re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 10, 2017 10:08

Hi Suarez!
i'm looking for DI-Tool software CD or program to connect would have it?

Log inor Register

to rate and post a comment

(331)
(67)

Gabriel Cool

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 10, 2017 04:47

Write me your email i will send you

1 person liked this
Log inor Register

to rate and post a comment

(10)
(1)

Sohail Aftab

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

March 16, 2017 11:06

sohail_aftab@live.com

Log inor Register

to rate and post a comment

teem teemm

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 14, 2017 06:05

Please, i need software for configuration lut dicom printer in drypix 7000.
Please E-mail me triptano@gmail.com

Log inor Register

to rate and post a comment

(331)
(67)

Gabriel Cool

re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 16, 2017 08:50

You can use M-utility on the printer, just push left upper (blue or green) buttom and then utility. I sended di tool for you, but you need set your PC as FTP server

Log inor Register

to rate and post a comment

(331)
(67)

Gabriel Cool

re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 16, 2017 08:51

Your email is wrong

Log inor Register

to rate and post a comment

teem teemm

re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

April 14, 2017 06:04

E-mail me software for triptano@gamil.com

Log inor Register

to rate and post a comment

(47)
(1)

Deniss Stalnov

re: re: re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

July 13, 2017 10:23

I have the same problem. Need help. Could you please send me Di-Tools too. denis.stalnov@gmail.com

Log inor Register

to rate and post a comment

(47)
(1)

Deniss Stalnov

re: fuji drypix 4000 or 7000 IP ADDRESS

July 13, 2017 10:22

I have the same problem. Need help.

Log inor Register

to rate and post a comment

(317)
(22)
(8)

Joseph Jenkins

re: re: fuji drypix 4000 or 7000 IP ADDRESS

August 14, 2018 04:28

please send me Di-Tools
iiltd@hotmail.com and service manual if possible thanks

Log inor Register

to rate and post a comment

(2)

(7)

Vladyslav Kyrychenko

re: re: re: fuji drypix 4000 or 7000 IP ADDRESS

September 06, 2018 09:42

Please send Di-Tools for Drypix 7000 to service@radio-med.com

Log inor Register

to rate and post a comment

(5)

sai

re: fuji drypix 4000 or 7000 IP ADDRESS

October 18, 2019 12:11

please send me, i need change IP addres DI TOOLS SOFTWEAR

Log inor Register

to rate and post a comment

stats
DOTmed text ad

stats
advertisement

Upcoming Auction Events

Houston Liquidator
TIMED
Mar 01, 2024 (551 lots)
Aga
TIMED
Mar 05, 2024 (12 lots)