GE VIVID 3 - Critical error

April 30, 2009 03:38


I need help with Vivid 3
Critical error when loading software, open the program and not open transducer. The external keyboard does not work.

thanks.

Log inor Register

to rate and post a comment

(46)
(16)

Ramil Cortes

Re: GE VIVID 3 - Critical error

May 01, 2009 10:46

kindly check your keyboard controller,try to reset or clean the said board..

Log inor Register

to rate and post a comment

(43)
(3)
(4)

Rodolfo Reyes Ingeniero

Re: GE VIVID 3 - Critical error

June 13, 2009 12:01

I have a similar trouble, the system not have the image of transducer.
Please help me, in this problem
Thank you

Log inor Register

to rate and post a comment

(56)
(3)
(1)

morteza yousefi

re: GE VIVID 3 - Critical error

October 11, 2013 08:43

this error some times is of beep or hi voltage.please chech the conection of the beep if is ok and you hi voltag is ok your kb controler is defectiv