Your Email:
Bruno Araujo
GE VIVID 3 - Critical error
April 30, 2009 03:38
I need help with Vivid 3 Critical error when loading software, open the program and not open transducer. The external keyboard does not work. thanks.
to rate and post a comment
Ramil Cortes
Re: GE VIVID 3 - Critical error
May 01, 2009 10:46
kindly check your keyboard controller,try to reset or clean the said board..
Rodolfo Reyes Ingeniero
June 13, 2009 12:01
I have a similar trouble, the system not have the image of transducer. Please help me, in this problem Thank you
morteza yousefi
re: GE VIVID 3 - Critical error
October 11, 2013 08:43
this error some times is of beep or hi voltage.please chech the conection of the beep if is ok and you hi voltag is ok your kb controler is defectiv
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Bruno Araujo
GE VIVID 3 - Critical error
April 30, 2009 03:38
I need help with Vivid 3
Critical error when loading software, open the program and not open transducer. The external keyboard does not work.
thanks.
to rate and post a comment
Ramil Cortes
Re: GE VIVID 3 - Critical error
May 01, 2009 10:46
kindly check your keyboard controller,try to reset or clean the said board..
to rate and post a comment
Rodolfo Reyes Ingeniero
Re: GE VIVID 3 - Critical error
June 13, 2009 12:01
I have a similar trouble, the system not have the image of transducer.
Please help me, in this problem
Thank you
to rate and post a comment
morteza yousefi
re: GE VIVID 3 - Critical error
October 11, 2013 08:43
this error some times is of beep or hi voltage.please chech the conection of the beep if is ok and you hi voltag is ok your kb controler is defectiv
to rate and post a comment