Over 500 Total Lots Up For Auction at One Location - NJ 06/15

help with trouble with LORAD MULTICARE PLATINUM

Tags: Other

Posted by an Anonymous User

help with trouble with LORAD MULTICARE PLATINUM

May 24, 2012 09:18

Some one help me, I have trouble with LORAD MULTICARE PLATINUM the table not up/down.
I check the test point on power control Board and all voltajes is good, but on TP6 not volatje.
Any have about this problem, please contact to me email> iseermed@gmail.com
Thank you for advance.

Log inor Register

to rate and post a comment

(6)
(1)

Brenda Fletcher

re: help with trouble with LORAD MULTICARE PLATINUM

May 29, 2012 11:13

Check the fuses behind the table.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)