Check the high voltage; if the high voltage is generated to less than 20 kV (check on D500_X186 at Pin UTact), then -> replace the tube unit, but first readout/note the tube data.
If no high voltage is generated, check the oscillation current signal on D500, Pin X210_Iload -> if the oscillation current signal is present, there is a short circuit in the HVT or in the high voltage circuit; for this, check for a short circuit on D530 X13 and X11 if there is no short circuit, then: - Replace tube unit, but first read out/note the tube data
Determine the oscillation with the single oscillation Test, check the control (D500_X100_F_INV) and oscillation pulse (D500_X210_I_load): - 1) If there is no control, then replace the D500. - 2) If no oscillation pulse is generated, replace the inverter
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Boris Janevski
Siemens Multix Swing Err 0042
November 02, 2023 10:56
Does anyone know how to troubleshot this error on Siemens Multix Swing Err 0042
Can someone please share Service manual for this modality?
Thank you in advance
to rate and post a comment
Niki
re: Siemens Multix Swing Err 0042
November 03, 2023 10:10
Hi, here is what you can try:
Check the high voltage; if the high voltage is generated to less than 20 kV (check on D500_X186 at Pin UTact), then -> replace the tube unit, but first readout/note the tube data.
If no high voltage is generated, check the oscillation current signal on D500, Pin X210_Iload -> if the oscillation current signal is present, there is a short circuit in the HVT or in the high voltage circuit; for this, check for a short circuit on D530 X13 and X11 if there is no short circuit, then: - Replace tube unit, but first read out/note the tube data
Determine the oscillation with the single oscillation Test, check the control (D500_X100_F_INV) and oscillation pulse (D500_X210_I_load): - 1) If there is no control, then replace the D500. - 2) If no oscillation pulse is generated, replace the inverter
BR,
Niki