Does anyone know how to troubleshot this error on Siemens Multix Swing Err 0042

Can someone please share Service manual for this modality?

Thank you in advance

Log inor Register

to rate and post a comment

Niki

re: Siemens Multix Swing Err 0042

November 03, 2023 10:10

Hi, here is what you can try:

Check the high voltage; if the high voltage is generated to less than 20 kV (check on D500_X186 at Pin UTact), then -> replace the tube unit, but first readout/note the tube data.
If no high voltage is generated, check the oscillation current signal on D500, Pin X210_Iload -> if the oscillation current signal is present, there is a short circuit in the HVT or in the high voltage circuit; for this, check for a short circuit on D530 X13 and X11 if there is no short circuit, then: - Replace tube unit, but first read out/note the tube data
Determine the oscillation with the single oscillation Test, check the control (D500_X100_F_INV) and oscillation pulse (D500_X210_I_load): - 1) If there is no control, then replace the D500. - 2) If no oscillation pulse is generated, replace the inverter

BR,
Niki

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)